Diagnostic Performance of an Electronic Nose, Fractional Exhaled Nitric Oxide,

and Lung Function Testing in Asthma

http://www.medscape.com/viewarticle/723250

From CHEST

Diagnostic Performance of an Electronic Nose, Fractional Exhaled Nitric Oxide,

and Lung Function Testing in Asthma

Posted: 08/19/2010; CHEST. 2010;137(4):790-796. © 2010 American College of Chest

Physicians

Abstract and Introduction

Abstract

Background: Analysis of exhaled breath by biosensors discriminates between

patients with asthma and healthy subjects. An electronic nose consists of a

chemical sensor array for the detection of volatile organic compounds (VOCs) and

an algorithm for pattern recognition. We compared the diagnostic performance of

a prototype of an electronic nose with lung function tests and fractional

exhaled nitric oxide (FENO) in patients with atopic asthma.

Methods: A cross-sectional study was undertaken in 27 patients with intermittent

and persistent mild asthma and in 24 healthy subjects. Two procedures for

collecting exhaled breath were followed to study the differences between total

and alveolar air. Seven patients with asthma and seven healthy subjects

participated in a study with mass spectrometry (MS) fingerprinting as an

independent technique for assessing between group discrimination. Classification

was based on principal component analysis and a feed-forward neural network.

Results: The best results were obtained when the electronic nose analysis was

performed on alveolar air. Diagnostic performance for electronic nose, FENO, and

lung function testing was 87.5%, 79.2%, and 70.8%, respectively. The combination

of electronic nose and FENO had the highest diagnostic performance for asthma

(95.8%). MS fingerprints of VOCs could discriminate between patients with asthma

and healthy subjects.

Conclusions: The electronic nose has a high diagnostic performance that can be

increased when combined with FENO. Large studies are now required to

definitively establish the diagnostic performance of the electronic nose.

Whether this integrated noninvasive approach will translate into an early

diagnosis of asthma has to be clarified.

Introduction

Several volatile organic compounds (VOCs) have been identified in exhaled breath

in healthy subjects by gas chromatography (GC)/mass spectrometry (MS).[1,2]

Identification of selective VOC patterns in exhaled breath is potentially useful

as a biomarker of asthma.[3] Differences between alveolar and

oropharyngeal/airway air can affect the results and should be considered when

analyzing VOCs in exhaled breath.[4] An electronic nose is an artificial sensor

system that consists of an array of chemical sensors for VOC detection and an

algorithm for pattern recognition.[5–7] The electronic nose discriminates

between patients with asthma and healthy subjects,[3] between patients with

asthma of different severity,[3] between patients with lung cancer and healthy

subjects,[8–10] and between patients with lung cancer and COPD.[9]

The primary aim of this study was to compare the diagnostic performance of a

prototype of an electronic nose with fractional exhaled nitric oxide (FENO), an

independent method for assessing airway inflammation, and lung function tests in

patients with asthma. Secondary aims of this study were to ascertain whether an

electronic nose could discriminate between patients with asthma and healthy

subjects and to establish the best sampling protocol (alveolar vs

oropharyngeal/airway air) for electronic nose analysis.

Materials and Methods

Study Subjects

Twenty-seven white patients with intermittent or mild persistent asthma and 24

healthy subjects were studied (Table 1). Among study subjects, seven patients

with asthma and seven healthy subjects participated in a study with GC/MS used

for MS fingerprinting, an independent technique for assessing between-group

discrimination.

[ CLOSE WINDOW ]

Table 1. Subject Characteristics

Characteristic Healthy Subjects Patients With Asthma

No. 24 27

Age, y 33 ± 3 39 ± 3

Sex, male (female) 12 (12) 12 (15)

FEV1, L 3.9 ± 0.2 3.4 ± 0.2a

FEV1, % predicted 108.2 ± 2.2 98.1 ± 2.5b

FVC, L 4.7 ± 1.3 4.5 ± 1.3

FVC, % predicted 109.2 ± 2.2 102.6 ± 4.6

FEF25%–75%, L 4.2 ± 1.2 3.1 ± 1.3b

FEF25%–75%, % 98.1 ± 5.0 75.3 ± 4.9b

Atopyc No Yes

Smoking history No No

Data are expressed as No. or mean ± SEM. FEF25%–75% = forced expiratory flow at

25% to 75% of forced vital capacity.

a P < .05 compared with healthy subjects.

b P < .01 compared with healthy subjects.

cPresence of atopy was confirmed by skin prick testing for common aeroallergens.

Patients with asthma had no corticosteroids or other antiinflammatory drugs for

asthma within 4 wk.

Patients with asthma were recruited from the Allergy Outpatient Clinic, Istituto

Dermopatico dell'Immacolata, IDI, Rome, Italy. Diagnosis and classification of

asthma were based on clinical history and examination and pulmonary function

parameters according to current guidelines.[11] Patients had intermittent asthma

with symptoms equal to or less often than twice a week (step 1) or mild

persistent asthma with symptoms more often than twice a week (step 2), FEV1 of

& #8805; 80% of predicted value and reversibility of & #8805; 12% to salbutamol,

or a positive provocation test result with methacholine or exercise. Patients

with asthma had atopy as confirmed by positive skin prick test results in

response to common aeroallergens. All atopic subjects had a clinical history of

atopy. They were not taking any regular medication, but used inhaled

short-acting & #946;2-agonists as needed for symptom relief. Healthy subjects had

no history of asthma or atopic disease, negative skin prick test results, and

normal spirometry. Subjects had no history of smoking, no upper respiratory

tract infections in the previous 3 weeks, and were excluded from the study if

they had used corticosteroids or antiinflammatory drugs for asthma in the

previous 4 weeks.

Study Design

The type of study was cross-sectional. Subjects, recruited from March 2007 to

September 2008, attended on one occasion for clinical examination, FENO

measurement, electronic nose analysis, lung function tests, and skin prick

testing. Informed consent was obtained from patients. The study was approved by

the Ethics Committee of the Catholic University of the Sacred Heart, Rome,

Italy.

Pulmonary Function and FENO Measurement

Spirometry was performed with a Pony FX spirometer (Cosmed; Rome, Italy). The

best of three consecutive maneuvers were chosen. FENO was measured with the NIOX

system (Aerocrine; Stockholm, Sweden) with a single-breath online method at

constant flow of 50 mL/s according to American Thoracic Society guidelines.[12]

FENO measurements were obtained before spirometry.

Collection of Exhaled Breath

Exhaled breath was collected from each subject at 8.30 AM. No food or drinks

were allowed at least 2 h prior to collection of exhaled breath. Two procedures

for collecting exhaled breath were followed to study the differences between

total exhaled breath and alveolar air and to establish the best protocol for the

electronic nose analysis. In the first sampling procedure (Fig 1A), subjects

were asked to inhale to total lung capacity and to exhale into a mouthpiece

connected to a Tedlar bag through a three-way valve. The total exhaled breath

was collected. In the second sampling procedure (Fig 1B), subjects were asked to

repeat the maneuver. The sampling system was designed considering a dead space

volume of 150 mL. The first 150 mL were collected into a separate Tedlar bag and

discarded, whereas the remaining exhaled breath, principally derived from the

alveolar compartment, was analyzed.

Figure 1. Exhaled breath sampling for electronic nose analysis. (A) Sampling of

total exhaled breath. (B) Sampling of alveolar exhaled air. The sampling system

was designed considering a dead space volume of 150 mL. The first part of

exhalation (150 mL) was discarded, whereas the remaining exhaled breath

principally derived from the alveolar compartment was collected for electronic

nose analysis.

Electronic Nose

A prototype electronic nose designed at the University of Rome Tor Vergata was

used.[13,14] The instrument contains an array of eight quartz microbalance gas

sensors coated by molecular films of metalloporphyrins.[15] Sensors detect the

amount of molecules absorbed in a sensitive film through the changes of resonant

frequency that is proportional to the absorbed mass.[15] The frequency shifts

are composed in patterns and analyzed by pattern recognition algorithms.[16]

Samples were analyzed immediately after their collection. Ambient VOCs were

subtracted from measures and results were automatically adjusted for ambient

VOCs.

GC/MS

GC/MS used for MS fingerprinting of VOCs[17] was performed (1) to confirm

between-group differences in VOC patterns detected with an electronic nose, and

(2) to ascertain whether exhaled breath samples were stable within 48 h from

collection. With MS fingerprinting, the sum of mass spectra of VOCs detected by

GC was considered as a sample pattern that was then analyzed like the electronic

nose patterns.[17] Total exhaled breath was collected (Fig 1A). Samples were

immediately absorbed onto sterile 10 × 10 cm2 gauze pads that were sealed in

20-mL headspace glass vials with crimped seal with

polytetrafluoroethylene/silicone septa (Supelco; Bellefonte, PA).[15] Samples

were put on ice, transferred to the laboratory, and stored at & #8722;10°C until

GC/MS analysis that was performed within 48 h from sample collection. Before

solid-phase microextraction, samples were kept at room temperature for 9 h. VOCs

were extracted from the vials using a divinylbenzene/carboxen on a

polydimethylsiloxane 50/30 m fiber (Supelco; Sigma-Aldrich; St. Louis, MO).

Exposure was performed at room temperature for 15 h by piercing the silicon

septum and inserting the fiber into the headspace of the vial. Solid-phase

microextraction fiber was then removed from the vial and transferred to the

injector of the gas chromatograph for thermal desorption.

VOCs adsorbed in the fiber were desorbed in the injection port of the GC for 3

min at an inlet temperature of 250°C in the splitless mode. Mass spectra were

obtained using electron ionization. The ionization occurred with a kinetic

energy of the impacting electrons of 70 eV.[18] Mass spectra and reconstructed

chromatograms (total ion current) were acquired in the full scan mode in the

mass range m/z 50–100.[15]

Skin Testing

Atopy was assessed by skin prick tests for common aeroallergens (mixture for

house dust mite [Dermatophagoides pteronyssinus and Dermatophagoides farinae],

grass pollen [cocksfoot and timothy], tree pollen [birch, ash tree, olive tree,

oak, and cypress], weed pollen [Ambrosia artemisifolia and Parietaria

officinalis], animal danders [cat and dog allergens], and fungal allergens

[Aspergillus species and Alternaria alternata]) (Stallergenes; Antony,

France).[19] A positive skin test response was defined as a wheal with a mean

diameter (mean of maximum and 90° midpoint diameters) of at least 3 mm greater

than that produced with a saline control.

Multivariate Data Analysis

The analysis of patterns requires multivariate statistical algorithms.[20] We

used principal component analysis to represent multidimensional data in a

bidimensional plane and feed-forward neural network to classify electronic nose,

FENO, and spirometry data. A feed-forward neural network is a biologically

derived classification model[21] that is formed by a number of processing units

(neurons), organized in layers. The total datasets have been divided in training

(27 measures) and testing (24 measures) set. To test the presence of any drift

in sensors, the first data collected were used for training and the remaining

data for testing.

Statistical Analysis

FENO values were expressed as medians and interquartile ranges (25th and 75th

percentiles). Spirometry values were expressed as mean ± SEM. Unpaired t test

and Mann-Whitney U test were used for comparing groups for normally distributed

and nonparametric data, respectively. Correlation was expressed as a Pearson

coefficient. Significance was defined as a value of P < .05.

Results

Electronic Nose

The best results were obtained when electronic nose analysis was performed on

alveolar air ( Table 2 and Table 3 ). Diagnostic classification with 95% CIs is

shown in Table 4 . The diagnostic performance was determined with the test

datasets in terms of the number of correct identifications of asthma diagnosis

based on current guidelines.[11] Diagnostic performance for the electronic nose,

FENO, lung function tests, and their combinations is shown in Table 2 and Table

3 and is related to the best performances obtained with the neural network for

each specific case. The combination of electronic nose analysis of alveolar air

and FENO had the highest diagnostic performance for asthma (95.8%). Electronic

nose analysis of alveolar air was able to discriminate between patients with

asthma and healthy subjects in 87.5% of cases, a diagnostic performance that was

higher than that of FENO (79.2%), spirometry (70.8%), and combination of FENO

and spirometry (83.3%). There was no correlation between the electronic nose,

FeNO, and lung function testing data in either the asthma or the healthy control

group.

[ CLOSE WINDOW ]

Table 2. Classification Matrices of the Feed-Forward Neural Network Classifier

in Training and Testing Phase for Data Related to Electronic Nose, Fractional

Exhaled Nitric Oxide, and Spirometry

Techniques Healthy Subjects Patients With Asthma

E-nose (alveolar exhaled air)

Training

Healthy subjects 11 1b

Patients with asthma 1a 14

Test

Healthy subjects 9 3b

Patients with asthma 0 12

E-nose (total exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 11 1b

Patients with asthma 5a 7

FENO

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 10 2b

Patients with asthma 3a 9

Spirometry

Training

Healthy subjects 11 1b

Patients with asthma 0 15

Test

Healthy subjects 9 3b

Patients with asthma 4a 8

E-nose analysis was performed in total exhaled air and alveolar exhaled air.

E-nose = electric nose; FENO = fractional exhaled nitric oxide.

aFalse negatives.

bFalse positives.

[ CLOSE WINDOW ]

Table 3. The Classification Matrices of the Feed-Forward Neural Network

Classifier in Training and Testing Phase for Data Related to Combination of

Electronic Nose, FENO, and Spirometry

Techniques Healthy Subjects Patients With Asthma

FENO and e-nose (alveolar exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 11 1b

Patients with asthma 0 12

FENO and e-nose (total exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 12 0

Patients with asthma 3a 9

Spirometry and e-nose (alveolar exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 9 3b

Patients with asthma 1a 11

Spirometry and e-nose (total exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 9 3b

Patients with asthma 1a 11

FENO and spirometry

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 10 2b

Patients with asthma 2a 10

Electronic nose analysis was performed in total exhaled air and alveolar exhaled

air. See Table 2 for expansion of abbreviations.

aFalse negatives.

bFalse positives.

[ CLOSE WINDOW ]

Table 4. Diagnostic Classification With 95% of CIs in Training and Testing Phase

for Data Related to Combination of Electronic Nose, FENO, and Spirometry

Technique Classification Rate, % Testing

E-nose (alveolar exhaled air) 92 86.2

E-nose (total exhaled air) 96 72.5

FENO 98.5 77.1

Spirometry 96 70.5

FENO and e-nose (alveolar exhaled air) 98.7 92.7

FENO and e-nose (total exhaled air) 99.1 86.6

Spirometry and e-nose (alveolar exhaled air) 99.3 81.8

FENO and spirometry 98.5 81.3

Spirometry and e-nose (total exhaled air) 99.1 81.1

Electronic nose analysis was performed in total exhaled air and alveolar exhaled

air. See Table 2 for expansion of abbreviations.

[ CLOSE WINDOW ]

Table 2. Classification Matrices of the Feed-Forward Neural Network Classifier

in Training and Testing Phase for Data Related to Electronic Nose, Fractional

Exhaled Nitric Oxide, and Spirometry

Techniques Healthy Subjects Patients With Asthma

E-nose (alveolar exhaled air)

Training

Healthy subjects 11 1b

Patients with asthma 1a 14

Test

Healthy subjects 9 3b

Patients with asthma 0 12

E-nose (total exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 11 1b

Patients with asthma 5a 7

FENO

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 10 2b

Patients with asthma 3a 9

Spirometry

Training

Healthy subjects 11 1b

Patients with asthma 0 15

Test

Healthy subjects 9 3b

Patients with asthma 4a 8

E-nose analysis was performed in total exhaled air and alveolar exhaled air.

E-nose = electric nose; FENO = fractional exhaled nitric oxide.

aFalse negatives.

bFalse positives.

[ CLOSE WINDOW ]

Table 3. The Classification Matrices of the Feed-Forward Neural Network

Classifier in Training and Testing Phase for Data Related to Combination of

Electronic Nose, FENO, and Spirometry

Techniques Healthy Subjects Patients With Asthma

FENO and e-nose (alveolar exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 11 1b

Patients with asthma 0 12

FENO and e-nose (total exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 12 0

Patients with asthma 3a 9

Spirometry and e-nose (alveolar exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 9 3b

Patients with asthma 1a 11

Spirometry and e-nose (total exhaled air)

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 9 3b

Patients with asthma 1a 11

FENO and spirometry

Training

Healthy subjects 12 0

Patients with asthma 0 15

Test

Healthy subjects 10 2b

Patients with asthma 2a 10

Electronic nose analysis was performed in total exhaled air and alveolar exhaled

air. See Table 2 for expansion of abbreviations.

aFalse negatives.

bFalse positives.

GC/MC

MS fingerprints of VOCs in exhaled breath in patients with asthma were different

from those in healthy subjects (Fig 2), confirming that the electronic nose used

in this study discriminates between patients with asthma and controls. MS

fingerprinting performed within 48 h from sampling was similar to that obtained

immediately after collection (data not shown), indicating that samples are

stable for at least 48 h.

Figure 2. PC analysis of mass spectrometry fingerprinting of patients with

asthma and healthy subjects. PC = principal component.

FENO and Lung Function Tests

Patients with asthma had higher median FENO values than healthy subjects (37. 6

[26.0–61.5] ppb vs 13.4 [10.0–19.9] ppb, P < .0001, respectively) (Fig 3).Both

study groups had normal FEV1 values (Table 1). Patients with asthma had lower

absolute (P = .032) and percentage of predicted FEV1 values (P = .004) than

controls (Table 1). Absolute (P = .003) and percentage of predicted forced

expiratory flow between 25% and 75% of forced vital capacity (FEF25%–75%) (P =

..002) were lower in patients with asthma than in healthy subjects (Table 1).

[ CLOSE WINDOW ]

Table 1. Subject Characteristics

Characteristic Healthy Subjects Patients With Asthma

No. 24 27

Age, y 33 ± 3 39 ± 3

Sex, male (female) 12 (12) 12 (15)

FEV1, L 3.9 ± 0.2 3.4 ± 0.2a

FEV1, % predicted 108.2 ± 2.2 98.1 ± 2.5b

FVC, L 4.7 ± 1.3 4.5 ± 1.3

FVC, % predicted 109.2 ± 2.2 102.6 ± 4.6

FEF25%–75%, L 4.2 ± 1.2 3.1 ± 1.3b

FEF25%–75%, % 98.1 ± 5.0 75.3 ± 4.9b

Atopyc No Yes

Smoking history No No

Data are expressed as No. or mean ± SEM. FEF25%–75% = forced expiratory flow at

25% to 75% of forced vital capacity.

a P < .05 compared with healthy subjects.

b P < .01 compared with healthy subjects.

cPresence of atopy was confirmed by skin prick testing for common aeroallergens.

Patients with asthma had no corticosteroids or other antiinflammatory drugs for

asthma within 4 wk.

[ CLOSE WINDOW ]

Table 1. Subject Characteristics

Characteristic Healthy Subjects Patients With Asthma

No. 24 27

Age, y 33 ± 3 39 ± 3

Sex, male (female) 12 (12) 12 (15)

FEV1, L 3.9 ± 0.2 3.4 ± 0.2a

FEV1, % predicted 108.2 ± 2.2 98.1 ± 2.5b

FVC, L 4.7 ± 1.3 4.5 ± 1.3

FVC, % predicted 109.2 ± 2.2 102.6 ± 4.6

FEF25%–75%, L 4.2 ± 1.2 3.1 ± 1.3b

FEF25%–75%, % 98.1 ± 5.0 75.3 ± 4.9b

Atopyc No Yes

Smoking history No No

Data are expressed as No. or mean ± SEM. FEF25%–75% = forced expiratory flow at

25% to 75% of forced vital capacity.

a P < .05 compared with healthy subjects.

b P < .01 compared with healthy subjects.

cPresence of atopy was confirmed by skin prick testing for common aeroallergens.

Patients with asthma had no corticosteroids or other antiinflammatory drugs for

asthma within 4 wk.

[ CLOSE WINDOW ]

Table 1. Subject Characteristics

Characteristic Healthy Subjects Patients With Asthma

No. 24 27

Age, y 33 ± 3 39 ± 3

Sex, male (female) 12 (12) 12 (15)

FEV1, L 3.9 ± 0.2 3.4 ± 0.2a

FEV1, % predicted 108.2 ± 2.2 98.1 ± 2.5b

FVC, L 4.7 ± 1.3 4.5 ± 1.3

FVC, % predicted 109.2 ± 2.2 102.6 ± 4.6

FEF25%–75%, L 4.2 ± 1.2 3.1 ± 1.3b

FEF25%–75%, % 98.1 ± 5.0 75.3 ± 4.9b

Atopyc No Yes

Smoking history No No

Data are expressed as No. or mean ± SEM. FEF25%–75% = forced expiratory flow at

25% to 75% of forced vital capacity.

a P < .05 compared with healthy subjects.

b P < .01 compared with healthy subjects.

cPresence of atopy was confirmed by skin prick testing for common aeroallergens.

Patients with asthma had no corticosteroids or other antiinflammatory drugs for

asthma within 4 wk.

Figure 3. FENO concentrations in 27 patients with asthma ( & #9642;) and 24

healthy subjects (•). Median values are shown with bars. FENO = fractional

exhaled nitric oxide.

Discussion

The original aspects of our study are: (1) the comparison between an electronic

nose and FENO, in addition to lung function tests; (2) the comparison between

total and alveolar exhaled air; (3) the number of study subjects (27 patients

with intermittent and persistent mild asthma and 24 healthy controls); (4) the

MS fingerprinting based on GC/MS analysis; and (5) the analysis of data based on

a neural network that included a training and test analysis performed in two

separate datasets for stringent quality control.

We compared the diagnostic performance of an electronic nose, FENO, and lung

function testing in patients with a physician-based diagnosis of asthma. The

combination of electronic nose and FENO had the best diagnostic performance for

asthma (95.7%). When alveolar exhaled air was sampled, the electronic nose had a

diagnostic performance of 87.5%, higher than that observed with FENO (79.2%),

spirometry (70.8%), and their combination (83.3%). Spirometry had the lowest

diagnostic performance in line with a previous prospective study that compared

FENO, sputum eosinophils, and FEV1 in children with asthma.[22] The present

study confirms that FENO has a good diagnostic performance for asthma. The

higher diagnostic performance of electronic nose observed in our study might

reflect the fact that electronic nose analyzes patterns generated by a complex

mixture of VOCs in exhaled breath, whereas the NO analyzer detects only one

biomolecule. A limitation of our study is the relatively small number of

subjects, which precludes definitive conclusions on asthma diagnostic

performance of electronic nose, FENO and lung function testing. Large-powered

studies are required to definitively establish the diagnostic performance of

these techniques in patients with a known diagnosis of asthma. We were unable to

ascertain whether the electronic nose can be used as a diagnostic tool for

screening of patients with asthma that requires large prospective studies.

MS fingerprinting confirms data obtained with the electronic nose and further

supports its use for asthma diagnosis. Identification and quantitative GC/MS

analysis of VOCs in exhaled breath are required to establish the

pathophysiological role of single VOCs in asthma. Taken together, these data

indicate that electronic nose might be useful for asthma diagnosis, particularly

in combination with FENO. Our results are consistent with a previous study

showing that a different electronic nose discriminates between patients with

asthma and healthy subjects[3] and prospect the possibility of using different

noninvasive techniques for achieving a greater asthma diagnostic performance.

Because asthma is principally characterized by airway inflammation, it may seem

surprising that the best results with the electronic nose were obtained when

collecting alveolar air rather than total exhaled breath that includes exhaled

breath from the airways. This might reflect the contribution of oropharyngeal

air to total breath.[4] A previous study with GC/MS has shown that for 47 VOCs

very significant differences in concentrations/detection were recorded between

alveolar and oropharyngeal air.[4] When sampling total breath, oropharyngeal air

is likely to introduce confounding factors that make electronic nose analysis

less reflective of what occurs in the respiratory system. Because of significant

interindividual variability in dead space volume, it is likely that mixed

airways/alveolar air rather than alveolar air was collected in our study. For

this reason, the results of electronic nose analysis of alveolar air could

partially reflect the production of VOCs within the peripheral airways.

The lack of correlation between electronic nose results and FENO might indicate

that these techniques reflect different aspects of airway inflammation. Formal

studies to compare electronic nose analysis with independent biomarkers of

airway inflammation (eg, FENO, sputum eosinophils)[12,23,24] and to ascertain

whether the electronic nose could be used for assessing and monitoring airway

inflammation in patients with asthma are warranted.

The electronic nose is not suitable for ascertaining the cellular source of VOCs

in exhaled breath that requires bronchoscopy studies. Chronic airway

inflammation can modify the metabolic pathways in patients with asthma as

reflected by increased concentrations of exhaled pentane, one of the end

products of lipid peroxidation.[25] Because patients included in our study were

not on regular antiinflammatory drugs for asthma, we were unable to assess the

effect of pharmacological therapy on VOC patterns in exhaled breath for which

controlled studies are required. Both the present and a previous study[3] aimed

at ascertaining whether electronic nose could discriminate between atopic

patients with asthma and healthy nonatopic subjects. For this reason, these

studies were unable to assess the contribution of atopy itself to the

classification based on the electronic nose. A control group of atopic subjects

without asthma should be included in future studies to clarify whether, and to

what extent, atopy is responsible for the selective VOC patterns in exhaled

breath observed in patients with atopic asthma.

When analyzing electronic nose data, validation of the statistical model used

for classification is essential. In contrast to a previous study in which the

same dataset was used for training the model and classification,[3] we used two

different datasets for training and testing that were obtained in different

periods of time, reducing the possibility of introducing a temporal bias in the

model. This way, the predictive capacity of the classification model is likely

to be more suitable for a real-life situation.

In conclusion, the electronic nose discriminates between patients with asthma

and healthy subjects and its performance is increased when combined with FENO.

Large studies are required to definitively establish the diagnostic performance

of the electronic nose. Whether this integrated noninvasive approach will

translate into an early diagnosis of asthma has to be clarified.

[ CLOSE WINDOW ]

References

1. M. Method for the collection and assay of volatile organic compounds

in breath. Anal Biochem 1997;247(2):272–278.

2. M, Herrera J, Krishnan S, Zain M, Greenberg J, Cataneo RN. Variation

in volatile organic compounds in the breath of normal humans. J Chromatogr B

Biomed Sci Appl 1999;729(1–2):75–88.

3.Dragonieri S, Schot R, Mertens BJ, et al. An electronic nose in the

discrimination of patients with asthma and controls. J Allergy Clin Immunol.

2007;120(4):856–862.

4.van den Velde S, Quirynen M, van Hee P, van Steenberghe D. Differences between

alveolar air and mouth air. Anal Chem 2007;79(9):3425–3429.

5. AP, Magan N. Electronic noses and disease diagnostics. Nat Rev

Microbiol 2004;2(2):161–166.

6.Röck F, Barsan N, Weimar U. Electronic nose: current status and future trends.

Chem Rev 2008;108(2):705–725.

7.D'Amico A, Di Natale C, Paolesse R, et al. Olfactory systems for medical

applications. Sens Actuators B Chem 2008;130(1):458–465.

8.Machado R, Laskowski D, Deffenderfer O, et al. Detection of lung cancer by

sensor array analyses of exhaled breath. Am J Respir Crit Care Med

2005;171(11):1286–1291.

9.Dragonieri S, Annema JT, Schot R, et al. An electronic nose in the

discrimination of patients with non-small cell lung cancer and COPD. Lung Cancer

2009;64(2):166–170.

10.Di Natale C, Macagnano A, elli E, et al. Lung cancer identification by

the analysis of breath by means of an array of non-selective gas sensors.

Biosens Bioelectron 2003;18(10):1209–1218.

11.National Asthma Education and Prevention Program. (2007) Expert Panel Report

III. Guidelines for the Diagnosis and Management of Asthma (National Heart,

Lung, and Blood Institute, Bethesda, MD) NIH publication 08–5847.

12.Recommendations for standardized procedures for the on-line and off-line

measurement of exhaled lower respiratory nitric oxide and nasal nitric oxide in

adults and children-1999. Am J Respir Crit Care Med 1999;160(6):2104–2117.

13.Di Natale C, Paolesse R, D'Amico A. Metalloporphyrins based artificial

olfactory receptors. Sens Actuators B Chem 2007;121(1):238–246.

14.D'Amico A, Di Natale C, Paolesse R, Mantini A, Macagnano A. Metalloporphyrins

as basic material for volatile sensitive sensors. Sens Actuators B Chem

2000;65(1–3):209–215.

15.D'Amico A, Bono R, Pennazza G, et al. Identification of melanoma with a gas

sensor array. Skin Res Technol 2008;14(2):226–236.

16.Gardner J, Bartlett P. A brief history of electronic noses. Sens Actuators B

Chem 1994;18(1–3):211–220.

17.Pavòn JL, M, Pinto C, Laespada E, Cordero M, Peña A. Strategies for

qualitative and quantitative analyses with mass spectrometry-based electronic

noses. Trends Analyt Chem 2006;25(3):257–266.

18.Kusch P, Knupp G. Headspace-SPME-GC-MS Identification of volatile organic

compounds released from expanded polystyrene. Journal of Polymers and the

Environment 2004;12(2):83–87.

19.Montuschi P, Mondino C, Koch P, PJ, Ciabattoni G. Effects of a

leukotriene receptor antagonist on exhaled leukotriene E4 and prostanoids in

children with asthma. J Allergy Clin Immunol 2006;118(2):347–353.

20. R, Wichern D. (1992) Applied Multivariate Statistical Analysis

(Prentice Hall, New York, NY), 3rd ed..

21.Bishop CM. (2006) Pattern Recognition and Machine Learning (Springer,

Heidelberg, Germany).

22.Sivan Y, Gadish T, Fireman E, Soferman R. The use of exhaled nitric oxide in

the diagnosis of asthma in school children. J Pediatr 2009;155(2):211–216.

23.Malerba M, Ragnoli B, Radaeli A, Tantucci C. Usefulness of exhaled nitric

oxide and sputum eosinophils in the long-term control of eosinophilic asthma.

Chest 2008;134(4):733–739.

24.Petsky HL, Kynaston JA, C, et al. Tailored interventions based on

sputum eosinophils versus clinical symptoms for asthma in children and adults.

Cochrane Database Syst Rev 2007;(2):CD005603.

25.Paredi P, Kharitonov SA, PJ. Elevation of exhaled ethane concentration

in asthma. Am J Respir Crit Care Med 2000;162(4 Pt 1):1450–1454.

Trial registration

EUDRACT https://eudralink.emea.europa.eu; Identifier: 2007-000890-51; and

clinicaltrials.gov; Identifier: NCT00819676.

Funding/Support

Supported by Merck, Sharp, and Dohme, and Catholic University of the Sacred

Heart academic grant 2008–2009.

Abbreviations

FEF25%–75% = forced expiratory flow at 25% to 75% of forced vital capacity; FENO

= fractional exhaled nitric oxide; GC = gas chromatography; MS = mass

spectrometry; VOC = volatile organic compound

Author contributions

Dr Montuschi: contributed to study planning, study design, measurement of FENO,

spirometry, data analysis, data interpretation, and wrote the manuscript.

Dr Santonico: contributed to electronic nose analysis, data interpretation, and

mass spectrometry fingerprinting.

Dr Mondino: contributed to recruitment of patients and skin prick testing.

Dr Pennazza: contributed to electronic nose analysis, data interpretation, and

mass spectrometry fingerprinting.

Ms Mantini: contributed to electronic nose analysis and mass spectrometry

fingerprinting.

Dr elli: contributed to multivariate data analysis and manuscript

preparation.

Ms Capuano: contributed to electronic nose analysis and mass spectrometry

fingerprinting.

Dr Ciabattoni: contributed to manuscript preparation and revision.

Dr Paolesse: contributed to biosensor manufacture.

Dr Di Natale: contributed to electronic nose setup, data interpretation, and

manuscript preparation.

Dr : contributed to manuscript preparation and revision.

Dr D'Amico: contributed to electronic nose setup, data interpretation, and

manuscript preparation and revision.

Financial/nonfinancial disclosures

The authors have reported to CHEST the following conflicts of interest: Dr

receives research funding from and has been on Scientific Advisory Boards

for AstraZeneca, Boehringer-Ingelheim, Chiesi, GlaxoKline, Novartis,

Pfizer, and UCB. Drs Montuschi, Santonico, Mondino, Pennazza, elli,

Ciabattoni, Paolesse, Di Natale, D'Amico, and Mss Mantini and Capuano have

reported to CHEST that no potential conflicts of interest exist with any

companies/organizations whose products or services may be discussed in this

article.

Other contributions

This work was performed at the Catholic University of the Sacred Heart, Rome,

Italy.

CHEST. 2010;137(4):790-796. © 2010 American College of Chest Physicians

Join the conversation

You are posting as a guest. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
×
×
  • Create New...