Guest guest Posted December 3, 2006 Report Share Posted December 3, 2006 Heck, you got me. Hmmm - I remember this being an issue on the earlier models of their product but, unless I slept through part of the inservice their company provided when we bought our first LP 12 at South EMS, I don't remember him mentioning the " diagnostic mode " issue with this version... So I would be interested to know the answer to this one too. Jane Hill -------------- Original message from " Donnie " : -------------- Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 3, 2006 Report Share Posted December 3, 2006 Donnie, I would call Mark Radford who is the Physio technical guy for our area, he should be able to provide you with what you need. Basically the issues is that the accuracy of the 12 lead, specifically the interpretation of ST segment analysis is not considered to be accurate unless it is performed and printed in diagnostic mode. Lee _____ From: texasems-l [mailto:texasems-l ] On Behalf Of Donnie Sent: Sunday, December 03, 2006 10:00 AM To: texasems-l Subject: LifePak 12 - Monitor Mode vs. Diagnostic Mode Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 3, 2006 Report Share Posted December 3, 2006 I emailed my medtronic rep and this is his reply " Mann, " wrote: Subject: Re: Fwd: LifePak 12 - Monitor Mode vs. Diagnostic Mode Date: Sun, 3 Dec 2006 10:18:04 -0600 To: Peggy, Thanks for this info. Can you forward this email to the person asking the question. First thing I would try is to replace the ekg cable, these wear out just like tires on a car. If that does not help then the 12 lead may need to be checked by our technicians and recalibrated. If they need something else have them call. Mann Hospital Sales Consultant Medtronic ERS keith.mann@... " It's not who we are underneath, but what we do that defines us. " --------------------------------- Need a quick answer? Get one in minutes from people who know. Ask your question on Yahoo! Answers. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 3, 2006 Report Share Posted December 3, 2006 Donnie, Some other things...first, the 12-lead has to be in diagnostic mode unless you are just doing it to kill little pink trees....(isn't that where they get the paper???) Secondly, how are our electrodes supplied? The ones we have used for years come 50 to a bag (10 strips of 5). One of the things we have discovered is that if we open that bag and we don't use them quickly (say in a day or so...depending upon heat and humidity) the gel is actually drying enough to cause messy base lines...so we keep them in ziplock bags...especially in our reserve units. Also, we always attempt to get our baseline 12-lead immediately upon contact with the patient...before going to the truck if possible. The truck moves too much (even not rolling) and there is way too much messy electricity leaking all over the place... One last thing if replacing the cables and keeping your electrodes sealed up doesn't work....where are your cell phones in comparison to the monitor and what phones are you using? I have seen some phones " warp " the monitor screen or almost completely turn it off if they are sitting on top of it when a call comes in....and the blackberry's and other smart phones are a ton worse....might be another place to look. Just don't shut diagnostic off... Dudley LifePak 12 - Monitor Mode vs. Diagnostic Mode Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone ________________________________________________________________________ Check out the new AOL. Most comprehensive set of free safety and security tools, free access to millions of high-quality videos from across the web, free AOL Mail and more. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 3, 2006 Report Share Posted December 3, 2006 The two biggest contributors to poor quality, diagnostic mode EKGs from Physio or any other manufacturer are electrodes and skin prep - not the machines. Electrodes - buy high quality, wet gel electrodes. Dry gel electrodes need time to " melt " into the skin. This becomes more of a problem during colder months. Dudley makes a good point about the quantity per package. Regardless of the quantity, work to keep them airtight and within an appropriate temperature range. Skin prep is a big killer of good tracings in diagnostic mode. Cleaning and shaving the skin as well as providing a light abrasion to the skin greatly enhances the likelihood of a clear tracing. Six years ago, our manufacturer spend thousands of dollars to evaluate everything about our machines (without saying we were at fault) - even flew in engineers to observe the machines in actual use. They bought new electrodes for us to try and provided new cables. Eventually, skin prep was the only thing left. Proper skin prep technique is vitally important. Checking the cables is another step. They do wear out with constant use. I have heard about bad lots although this is reportedly rare. Electrode placement and patient instructions may impact diagnostic quality tracings. You might have a hospital 12 lead tech observe your practice to seek tips on subtle changes to improve technique. Obtaining clear tracings in diagnostic mode requires perfect technique, optimal conditions, and finesse. Our environment rarely promotes all three. ************************************************ , MPA, LP Chief Administrative Officer Montgomery County Hospital District ________________________________ From: texasems-l [mailto:texasems-l ] On Behalf Of Donnie Sent: Sunday, December 03, 2006 10:00 AM To: texasems-l Subject: LifePak 12 - Monitor Mode vs. Diagnostic Mode Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone *************************************************************************** The contents of this communication are intended only for the addressee and may contain confidential and/or privileged material. If you are not the intended recipient, please do not read, copy, use or disclose this communication and notify the sender. Opinions, conclusions and other information in this communication that do not relate to the official business of Montgomery County Hospital District shall be understood as neither given nor endorsed by it. *************************************************************************** Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 Peggy, but he doesn't mention anything in his e-mail response about diagnostic mode. This is the same fellow who taught our inservice and I don't recall him mentioning it there either. Can you get back with him and clarify his e-mail response? Jane Hill --------- Re: Fwd: LifePak 12 - Monitor Mode vs. Diagnostic Mode Date: Sun, 3 Dec 2006 10:18:04 -0600 To: Peggy, Thanks for this info. Can you forward this email to the person asking the question. First thing I would try is to replace the ekg cable, these wear out just like tires on a car. If that does not help then the 12 lead may need to be checked by our technicians and recalibrated. If they need something else have them call. Mann Hospital Sales Consultant Medtronic ERS keith.mann@... " It's not who we are underneath, but what we do that defines us. " --------------------------------- Need a quick answer? Get one in minutes from people who know. Ask your question on Yahoo! Answers. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 When I worked in Rockport, we had the same issue while in diagnostic mode. Physio was quick to try and isolate the problem and send us new cables, didn't help We eventually tracked it down to older electrodes, even after opening new bags from our own supply, we still had the same problem, we started ordering electrodes in smaller quantities, and smaller bags, cured the problem. Ours had become somewhat deteriorated, and we would get a constant interference regardless of where we were. I would honestly start there. W. Hatfield FF/EMT-P www.canyonlakefire-ems.org " Ubi concordia, ibi victoria " Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode Peggy, but he doesn't mention anything in his e-mail response about diagnostic mode. This is the same fellow who taught our inservice and I don't recall him mentioning it there either. Can you get back with him and clarify his e-mail response? Jane Hill --------- Re: Fwd: LifePak 12 - Monitor Mode vs. Diagnostic Mode Date: Sun, 3 Dec 2006 10:18:04 -0600 From: " Mann, " <keith.mann@medtroni <mailto:keith.mann%40medtronic.com> c.com> To: <stinnett_ems@ <mailto:stinnett_ems%40yahoo.com> yahoo.com> Peggy, Thanks for this info. Can you forward this email to the person asking the question. First thing I would try is to replace the ekg cable, these wear out just like tires on a car. If that does not help then the 12 lead may need to be checked by our technicians and recalibrated. If they need something else have them call. Mann Hospital Sales Consultant Medtronic ERS keith.mann@medtroni <mailto:keith.mann%40medtronic.com> c.com " It's not who we are underneath, but what we do that defines us. " --------------------------------- Need a quick answer? Get one in minutes from people who know. Ask your question on Yahoo! Answers. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 Donnie, I am certain you have tried it, and seen it suggested but I will for the sake of putting in my 2cents say it again. We have 24 LP12's, our biggest issue has been employees putting the electrodes on before use and placement of the leads while " hanging around " being right where the AC/Heat blows. We told our employees to look at the package and read where it says " do not open until patient use " and if there is a problem with the interpretation or the baseline to open a new package. This has solved all but the old cables and the employee's who have taped the lead wires together (LL and RL) and (RA and LA). Ok, I feel better now, Chris Donnie wrote: Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone --------------------------------- Want to start your own business? Learn how on Yahoo! Small Business. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 You are gonna laught at this but I recently had one of those, " Wow, I could have had a V8 " moments when I had difficulty getting a clean 12 lead reading on a patient. It just looked like gobbledygook each time I ran it - like we were bouncing down the road instead of sitting still. Then my student and I looked at each other as the lightbulb came on in both of our heads instantaneously when we noticed that we had placed his aluminum cane on the stretcher with him. The leads were lying across the cane....... We removed the cane, and WALA - a perfect 12 lead strip. LOL Jane Hill -------------- Original message from Weinzapfel : -------------- Donnie, I am certain you have tried it, and seen it suggested but I will for the sake of putting in my 2cents say it again. We have 24 LP12's, our biggest issue has been employees putting the electrodes on before use and placement of the leads while " hanging around " being right where the AC/Heat blows. We told our employees to look at the package and read where it says " do not open until patient use " and if there is a problem with the interpretation or the baseline to open a new package. This has solved all but the old cables and the employee's who have taped the lead wires together (LL and RL) and (RA and LA). Ok, I feel better now, Chris Donnie wrote: Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone --------------------------------- Want to start your own business? Learn how on Yahoo! Small Business. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 Remember also, that those of us who are so proud (previous posts) of our Frazer ambulances need to remember that the LP12 cannot sit on the bench seat in most cases, and get a decent tracing due to the magnetic field the generator coils create...even when it is not running. Any of you using Alaris Med III pumps may also need to relocate the pump and/or monitor, as that is a known cause of artifact. Last but not least, last month while traveling down the road, lightening struck very, very near the unit and the LP 12 went completely haywire. I had to power it down, and back up, so it would reboot, before it would do anything correctly. Makes me wonder how many " little " glitches we experience routinely that are attributable to storm and static activity. I know the lightning was not aimed for us...after all we EMS types are all the " clean living " types, right! Vernon Gresham City of Ganado Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode You are gonna laught at this but I recently had one of those, " Wow, I could have had a V8 " moments when I had difficulty getting a clean 12 lead reading on a patient. It just looked like gobbledygook each time I ran it - like we were bouncing down the road instead of sitting still. Then my student and I looked at each other as the lightbulb came on in both of our heads instantaneously when we noticed that we had placed his aluminum cane on the stretcher with him. The leads were lying across the cane....... We removed the cane, and WALA - a perfect 12 lead strip. LOL Jane Hill -------------- Original message from Weinzapfel : -------------- Donnie, I am certain you have tried it, and seen it suggested but I will for the sake of putting in my 2cents say it again. We have 24 LP12's, our biggest issue has been employees putting the electrodes on before use and placement of the leads while " hanging around " being right where the AC/Heat blows. We told our employees to look at the package and read where it says " do not open until patient use " and if there is a problem with the interpretation or the baseline to open a new package. This has solved all but the old cables and the employee's who have taped the lead wires together (LL and RL) and (RA and LA). Ok, I feel better now, Chris Donnie wrote: Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone --------------------------------- Want to start your own business? Learn how on Yahoo! Small Business. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 Since there are various designs on Frazer ambulances, I am not sure I follow. On the unit I helped purchase, the generator was on the opposite side of the box from the bench seat. Am I just being a redhead?? LOL Jane Hill --------- Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode You are gonna laught at this but I recently had one of those, " Wow, I could have had a V8 " moments when I had difficulty getting a clean 12 lead reading on a patient. It just looked like gobbledygook each time I ran it - like we were bouncing down the road instead of sitting still. Then my student and I looked at each other as the lightbulb came on in both of our heads instantaneously when we noticed that we had placed his aluminum cane on the stretcher with him. The leads were lying across the cane....... We removed the cane, and WALA - a perfect 12 lead strip. LOL Jane Hill -------------- Original message from Weinzapfel <ctacdoc657 (AT) yahoo (DOT) <mailto:ctacdoc657%40yahoo.com> com>: -------------- Donnie, I am certain you have tried it, and seen it suggested but I will for the sake of putting in my 2cents say it again. We have 24 LP12's, our biggest issue has been employees putting the electrodes on before use and placement of the leads while " hanging around " being right where the AC/Heat blows. We told our employees to look at the package and read where it says " do not open until patient use " and if there is a problem with the interpretation or the baseline to open a new package. This has solved all but the old cables and the employee's who have taped the lead wires together (LL and RL) and (RA and LA). Ok, I feel better now, Chris Donnie <wdstone (AT) sbcglobal (DOT) <mailto:wdstone%40sbcglobal.net> net> wrote: Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone --------------------------------- Want to start your own business? Learn how on Yahoo! Small Business. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 I had heard about the Frazer issue in the past, but have never had an issue with it myself, and have run countless 12 leads while the LP12 was on the bench. Interesting piece of information though. W. Hatfield FF/EMT-P www.canyonlakefire-ems.org " Ubi concordia, ibi victoria " Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode Remember also, that those of us who are so proud (previous posts) of our Frazer ambulances need to remember that the LP12 cannot sit on the bench seat in most cases, and get a decent tracing due to the magnetic field the generator coils create...even when it is not running. Any of you using Alaris Med III pumps may also need to relocate the pump and/or monitor, as that is a known cause of artifact. Last but not least, last month while traveling down the road, lightening struck very, very near the unit and the LP 12 went completely haywire. I had to power it down, and back up, so it would reboot, before it would do anything correctly. Makes me wonder how many " little " glitches we experience routinely that are attributable to storm and static activity. I know the lightning was not aimed for us...after all we EMS types are all the " clean living " types, right! Vernon Gresham City of Ganado Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode You are gonna laught at this but I recently had one of those, " Wow, I could have had a V8 " moments when I had difficulty getting a clean 12 lead reading on a patient. It just looked like gobbledygook each time I ran it - like we were bouncing down the road instead of sitting still. Then my student and I looked at each other as the lightbulb came on in both of our heads instantaneously when we noticed that we had placed his aluminum cane on the stretcher with him. The leads were lying across the cane....... We removed the cane, and WALA - a perfect 12 lead strip. LOL Jane Hill -------------- Original message from Weinzapfel <ctacdoc657 (AT) yahoo (DOT) <mailto:ctacdoc657%40yahoo.com> com>: -------------- Donnie, I am certain you have tried it, and seen it suggested but I will for the sake of putting in my 2cents say it again. We have 24 LP12's, our biggest issue has been employees putting the electrodes on before use and placement of the leads while " hanging around " being right where the AC/Heat blows. We told our employees to look at the package and read where it says " do not open until patient use " and if there is a problem with the interpretation or the baseline to open a new package. This has solved all but the old cables and the employee's who have taped the lead wires together (LL and RL) and (RA and LA). Ok, I feel better now, Chris Donnie <wdstone (AT) sbcglobal (DOT) <mailto:wdstone%40sbcglobal.net> net> wrote: Hey folks, I need some information regarding the use of diagnostic mode vs. monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have tried to explain to the " higher ups " why we must use diagnostic mode; however, I am having difficulties in locating any specific, tangible evidence to back this up. The reason there is an issue is that when we do a 12 lead in diagnostic mode (which is the current setting), the majority of the time there is so much interference that it can not be read. FYI: This is somewhat of a new problem and we haven't changed our electrodes. Could this be an issue with the machines themselves? Can anyone help me out with this? Any help would be greatly appreciated! D. Stone --------------------------------- Want to start your own business? Learn how on Yahoo! Small Business. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 First of all, thanks for all of your input. Secondly, after much discussion, we have narrowed it down to an issue that we will be working on. We currently have electrodes on the cables at all times. So, we are looking at changing this and only getting the electrodes out per individual need. Also, about 6 months ago, we began ordering our electrodes in bulk (30 per package). This means that each time a new package was opened the ones that were used last could have easily dried out prior to use. We currently use the Silvon Diaphoretic electrodes and they seem to only have a minimal amount of conductive get to begin with. We will be hitting the electrode issue hard to see if this will correct our problems. After a discussion with our Physio rep, he stated that the diagnostic mode was more for hospital use and not EMS. This is something that I totally do not agree with as the diagnostic mode is the only true way to zero in on the j-point and ST segment for analysis of changes from the baseline (as beautifully stated by Lee . I had to give him the credit for that statement). Monitor mode only zeros in on the QRS to give the rate and rhythm. Lastly, we also have Frazer ambulances, but have only had this issue for the last 6 months or so. Seems to be around the same time we changed the way the electrodes are packaged. Thanks again for all of your input!!! D. Stone > Hey folks, > > I need some information regarding the use of diagnostic mode vs. > monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have > tried to explain to the " higher ups " why we must use diagnostic mode; > however, I am having difficulties in locating any specific, tangible > evidence to back this up. The reason there is an issue is that when we > do a 12 lead in diagnostic mode (which is the current setting), the > majority of the time there is so much interference that it can not be > read. FYI: This is somewhat of a new problem and we haven't changed > our electrodes. Could this be an issue with the machines themselves? > Can anyone help me out with this? Any help would be greatly > appreciated! > > D. Stone > > --------------------------------- > Want to start your own business? Learn how on Yahoo! Small Business. > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 4, 2006 Report Share Posted December 4, 2006 We use the Blue Dot SP-00S electrodes with great success in diagnostic mode. They are packaged in 30 count packages but have a great deal of conductive gel on them to begin with. We also keep electrodes on the cables at all times and have not had an issue with them drying out. Then again the average time the electrodes stay on the cables without use is less than 24 hours at the longest. These electrodes also tend to be more adhesive with diaphoretic pts than those we used before. I would recommend these to anyone but they do tend to cost a little bit more than our previous ones but the results are well worth the money. Ince From: texasems-l [mailto:texasems-l ] On Behalf Of Donnie Sent: Monday, December 04, 2006 7:32 PM To: texasems-l Subject: Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode First of all, thanks for all of your input. Secondly, after much discussion, we have narrowed it down to an issue that we will be working on. We currently have electrodes on the cables at all times. So, we are looking at changing this and only getting the electrodes out per individual need. Also, about 6 months ago, we began ordering our electrodes in bulk (30 per package). This means that each time a new package was opened the ones that were used last could have easily dried out prior to use. We currently use the Silvon Diaphoretic electrodes and they seem to only have a minimal amount of conductive get to begin with. We will be hitting the electrode issue hard to see if this will correct our problems. After a discussion with our Physio rep, he stated that the diagnostic mode was more for hospital use and not EMS. This is something that I totally do not agree with as the diagnostic mode is the only true way to zero in on the j-point and ST segment for analysis of changes from the baseline (as beautifully stated by Lee . I had to give him the credit for that statement). Monitor mode only zeros in on the QRS to give the rate and rhythm. Lastly, we also have Frazer ambulances, but have only had this issue for the last 6 months or so. Seems to be around the same time we changed the way the electrodes are packaged. Thanks again for all of your input!!! D. Stone > Hey folks, > > I need some information regarding the use of diagnostic mode vs. > monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have > tried to explain to the " higher ups " why we must use diagnostic mode; > however, I am having difficulties in locating any specific, tangible > evidence to back this up. The reason there is an issue is that when we > do a 12 lead in diagnostic mode (which is the current setting), the > majority of the time there is so much interference that it can not be > read. FYI: This is somewhat of a new problem and we haven't changed > our electrodes. Could this be an issue with the machines themselves? > Can anyone help me out with this? Any help would be greatly > appreciated! > > D. Stone > > --------------------------------- > Want to start your own business? Learn how on Yahoo! Small Business. > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 5, 2006 Report Share Posted December 5, 2006 Maybe that is why the rep who did our inservice on these things didn't even mention diagnostic mode? They think we don't need it or use it. Ok, since it wasn't included in OUR inservice, how do you preset the LP 12 in diagnostic mode so it runs there and, if there is a problem, how do you get it back into regular mode? Jane -------------- Original message from " Donnie " : -------------- First of all, thanks for all of your input. Secondly, after much discussion, we have narrowed it down to an issue that we will be working on. We currently have electrodes on the cables at all times. So, we are looking at changing this and only getting the electrodes out per individual need. Also, about 6 months ago, we began ordering our electrodes in bulk (30 per package). This means that each time a new package was opened the ones that were used last could have easily dried out prior to use. We currently use the Silvon Diaphoretic electrodes and they seem to only have a minimal amount of conductive get to begin with. We will be hitting the electrode issue hard to see if this will correct our problems. After a discussion with our Physio rep, he stated that the diagnostic mode was more for hospital use and not EMS. This is something that I totally do not agree with as the diagnostic mode is the only true way to zero in on the j-point and ST segment for analysis of changes from the baseline (as beautifully stated by Lee . I had to give him the credit for that statement). Monitor mode only zeros in on the QRS to give the rate and rhythm. Lastly, we also have Frazer ambulances, but have only had this issue for the last 6 months or so. Seems to be around the same time we changed the way the electrodes are packaged. Thanks again for all of your input!!! D. Stone > Hey folks, > > I need some information regarding the use of diagnostic mode vs. > monitor mode while obtaining 12 lead EKGs in the LifePak 12. I have > tried to explain to the " higher ups " why we must use diagnostic mode; > however, I am having difficulties in locating any specific, tangible > evidence to back this up. The reason there is an issue is that when we > do a 12 lead in diagnostic mode (which is the current setting), the > majority of the time there is so much interference that it can not be > read. FYI: This is somewhat of a new problem and we haven't changed > our electrodes. Could this be an issue with the machines themselves? > Can anyone help me out with this? Any help would be greatly > appreciated! > > D. Stone > > --------------------------------- > Want to start your own business? Learn how on Yahoo! Small Business. > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 5, 2006 Report Share Posted December 5, 2006 another thing which will cause EMI to the monitor is the fluorescent lights in the ambulance. the ecg monitors are in a plastic-like case and are ungrounded jim davis Hatfield wrote: When I worked in Rockport, we had the same issue while in diagnostic mode. Physio was quick to try and isolate the problem and send us new cables, didn't help We eventually tracked it down to older electrodes, even after opening new bags from our own supply, we still had the same problem, we started ordering electrodes in smaller quantities, and smaller bags, cured the problem. Ours had become somewhat deteriorated, and we would get a constant interference regardless of where we were. I would honestly start there. W. Hatfield FF/EMT-P www.canyonlakefire-ems.org " Ubi concordia, ibi victoria " Re: LifePak 12 - Monitor Mode vs. Diagnostic Mode Peggy, but he doesn't mention anything in his e-mail response about diagnostic mode. This is the same fellow who taught our inservice and I don't recall him mentioning it there either. Can you get back with him and clarify his e-mail response? Jane Hill --------- Re: Fwd: LifePak 12 - Monitor Mode vs. Diagnostic Mode Date: Sun, 3 Dec 2006 10:18:04 -0600 From: " Mann, " <keith.mann@medtroni <mailto:keith.mann%40medtronic.com> c.com> To: <stinnett_ems@ <mailto:stinnett_ems%40yahoo.com> yahoo.com> Peggy, Thanks for this info. Can you forward this email to the person asking the question. First thing I would try is to replace the ekg cable, these wear out just like tires on a car. If that does not help then the 12 lead may need to be checked by our technicians and recalibrated. If they need something else have them call. Mann Hospital Sales Consultant Medtronic ERS keith.mann@medtroni <mailto:keith.mann%40medtronic.com> c.com " It's not who we are underneath, but what we do that defines us. " --------------------------------- Need a quick answer? Get one in minutes from people who know. Ask your question on Yahoo! Answers. Quote Link to comment Share on other sites More sharing options...