The 550 filter is shallow penetration and concentrates heat at the surface

layers of the skin, so it generally tends to cause more redness and post op

flareups than the 570 and 590, which penetrate deeper and carry heat with

them. The Epilight has a three second pause between blasts as opposed to the

Photoderm's 10 second pause, and the Epilight energy settings are roughly 10

points higher in impact than Photoderm. That is to say that 50j on Epilight

equals 60j on Photoderm. The Epilight will handle a 590 filter but not a

550. Not sure about the 570. ***Someone please correct me if I'm wrong on

any of the above.***

Here are my settings for tx #5 on 5/26/00:

Device=Epilight

Filter=590

Fluence=52j

Pulses#=3

Pulse Duration=3.0ms

Pulse Delay=20.0ms

Here are my settings for tx #9 on 5/26/00:

This tx was more aggressive than #5.

Device=Epilight

Filter=590

Fluence=60j

Pulses#=3

Pulse Duration=3.0ms

Pulse Delay=10.0ms

Notice that the joules were higher on tx 9, plus the delay between pulses

was shortened from 20 to 10 ms. The shortened pulse delay was intended to

cause more heat retention by targeted vessels between pulses. Also I

received over 450 blasts (1350+ pulses) during tx 9, as compared to about

320 blasts for tx 5. Please keep in mind that joule settings on Epilight are

more aggressive than the Photoderm equivalent. Your doc should know this but

you might want to query him if he decides to use the Epilight. Hope this

helps.

~Jim

_________________________________________________________________

Get your FREE download of MSN Explorer at http://explorer.msn.com

Link to comment
Share on other sites

Sheila,

The 550 filter is shallow penetration and concentrates heat at the surface

layers of the skin, so it generally tends to cause more redness and post op

flareups than the 570 and 590, which penetrate deeper and carry heat with

them. The Epilight has a three second pause between blasts as opposed to the

Photoderm's 10 second pause, and the Epilight energy settings are roughly 10

points higher in impact than Photoderm. That is to say that 50j on Epilight

equals 60j on Photoderm. The Epilight will handle a 590 filter but not a

550. Not sure about the 570. ***Someone please correct me if I'm wrong on

any of the above.***

Here are my settings for tx #5 on 5/26/00:

Device=Epilight

Filter=590

Fluence=52j

Pulses#=3

Pulse Duration=3.0ms

Pulse Delay=20.0ms

Here are my settings for tx #9 on 5/26/00:

This tx was more aggressive than #5.

Device=Epilight

Filter=590

Fluence=60j

Pulses#=3

Pulse Duration=3.0ms

Pulse Delay=10.0ms

Notice that the joules were higher on tx 9, plus the delay between pulses

was shortened from 20 to 10 ms. The shortened pulse delay was intended to

cause more heat retention by targeted vessels between pulses. Also I

received over 450 blasts (1350+ pulses) during tx 9, as compared to about

320 blasts for tx 5. Please keep in mind that joule settings on Epilight are

more aggressive than the Photoderm equivalent. Your doc should know this but

you might want to query him if he decides to use the Epilight. Hope this

helps.

~Jim

_________________________________________________________________

Get your FREE download of MSN Explorer at http://explorer.msn.com

Link to comment
Share on other sites

Jim, Sheila,

My 4th treatment used following parameters:

Device=Epilight

Filter=590

Fluence=48j

Pulses#=3

Pulse Duration=3.5ms

Pulse Delay=10.0ms

Note that these are quite similar to what Jim received. Perhaps not

surprising, since we are both being treated by Bitter Sr trained

derms with Epilight machine.

Jim is absolutely right on difference between machines, but I have

recently learned that individual machines 'drift' and periodically

need to be recalibrated to maintain same fluence - hence actual

delivered fluences can vary between two identical machines ... (as

if this wasn't already complicated enough ...)

I started to see my very first really noticeable improvement in both

facial telangiectasia and background redness about 2.5 weeks after

this 4th treatment. BTW, telangiectasia were initially passed with

570 filter, and then the above settings were applied to full face.

Rick

> Sheila,

> The 550 filter is shallow penetration and concentrates heat at the

surface

> layers of the skin, so it generally tends to cause more redness and

post op

> flareups than the 570 and 590, which penetrate deeper and carry

heat

with

> them. The Epilight has a three second pause between blasts as

opposed to the

> Photoderm's 10 second pause, and the Epilight energy settings are

roughly 10

> points higher in impact than Photoderm. That is to say that 50j on

Epilight

> equals 60j on Photoderm. The Epilight will handle a 590 filter but

not a

> 550. Not sure about the 570. ***Someone please correct me if I'm

wrong on

> any of the above.***

>

> Here are my settings for tx #5 on 5/26/00:

> Device=Epilight

> Filter=590

> Fluence=52j

> Pulses#=3

> Pulse Duration=3.0ms

> Pulse Delay=20.0ms

>

> Here are my settings for tx #9 on 5/26/00:

> This tx was more aggressive than #5.

> Device=Epilight

> Filter=590

> Fluence=60j

> Pulses#=3

> Pulse Duration=3.0ms

> Pulse Delay=10.0ms

>

> Notice that the joules were higher on tx 9, plus the delay between

pulses

> was shortened from 20 to 10 ms. The shortened pulse delay was

intended to

> cause more heat retention by targeted vessels between pulses. Also

I

> received over 450 blasts (1350+ pulses) during tx 9, as compared to

about

> 320 blasts for tx 5. Please keep in mind that joule settings on

Epilight are

> more aggressive than the Photoderm equivalent. Your doc should know

this but

> you might want to query him if he decides to use the Epilight. Hope

this

> helps.

> ~Jim

>

>

> _________________________________________________________________

> Get your FREE download of MSN Explorer at http://explorer.msn.com

Link to comment
Share on other sites

Jim, Sheila,

My 4th treatment used following parameters:

Device=Epilight

Filter=590

Fluence=48j

Pulses#=3

Pulse Duration=3.5ms

Pulse Delay=10.0ms

Note that these are quite similar to what Jim received. Perhaps not

surprising, since we are both being treated by Bitter Sr trained

derms with Epilight machine.

Jim is absolutely right on difference between machines, but I have

recently learned that individual machines 'drift' and periodically

need to be recalibrated to maintain same fluence - hence actual

delivered fluences can vary between two identical machines ... (as

if this wasn't already complicated enough ...)

I started to see my very first really noticeable improvement in both

facial telangiectasia and background redness about 2.5 weeks after

this 4th treatment. BTW, telangiectasia were initially passed with

570 filter, and then the above settings were applied to full face.

Rick

> Sheila,

> The 550 filter is shallow penetration and concentrates heat at the

surface

> layers of the skin, so it generally tends to cause more redness and

post op

> flareups than the 570 and 590, which penetrate deeper and carry

heat

with

> them. The Epilight has a three second pause between blasts as

opposed to the

> Photoderm's 10 second pause, and the Epilight energy settings are

roughly 10

> points higher in impact than Photoderm. That is to say that 50j on

Epilight

> equals 60j on Photoderm. The Epilight will handle a 590 filter but

not a

> 550. Not sure about the 570. ***Someone please correct me if I'm

wrong on

> any of the above.***

>

> Here are my settings for tx #5 on 5/26/00:

> Device=Epilight

> Filter=590

> Fluence=52j

> Pulses#=3

> Pulse Duration=3.0ms

> Pulse Delay=20.0ms

>

> Here are my settings for tx #9 on 5/26/00:

> This tx was more aggressive than #5.

> Device=Epilight

> Filter=590

> Fluence=60j

> Pulses#=3

> Pulse Duration=3.0ms

> Pulse Delay=10.0ms

>

> Notice that the joules were higher on tx 9, plus the delay between

pulses

> was shortened from 20 to 10 ms. The shortened pulse delay was

intended to

> cause more heat retention by targeted vessels between pulses. Also

I

> received over 450 blasts (1350+ pulses) during tx 9, as compared to

about

> 320 blasts for tx 5. Please keep in mind that joule settings on

Epilight are

> more aggressive than the Photoderm equivalent. Your doc should know

this but

> you might want to query him if he decides to use the Epilight. Hope

this

> helps.

> ~Jim

>

>

> _________________________________________________________________

> Get your FREE download of MSN Explorer at http://explorer.msn.com

Link to comment
Share on other sites

Jim, Sheila,

My 4th treatment used following parameters:

Device=Epilight

Filter=590

Fluence=48j

Pulses#=3

Pulse Duration=3.5ms

Pulse Delay=10.0ms

Note that these are quite similar to what Jim received. Perhaps not

surprising, since we are both being treated by Bitter Sr trained

derms with Epilight machine.

Jim is absolutely right on difference between machines, but I have

recently learned that individual machines 'drift' and periodically

need to be recalibrated to maintain same fluence - hence actual

delivered fluences can vary between two identical machines ... (as

if this wasn't already complicated enough ...)

I started to see my very first really noticeable improvement in both

facial telangiectasia and background redness about 2.5 weeks after

this 4th treatment. BTW, telangiectasia were initially passed with

570 filter, and then the above settings were applied to full face.

Rick

> Sheila,

> The 550 filter is shallow penetration and concentrates heat at the

surface

> layers of the skin, so it generally tends to cause more redness and

post op

> flareups than the 570 and 590, which penetrate deeper and carry

heat

with

> them. The Epilight has a three second pause between blasts as

opposed to the

> Photoderm's 10 second pause, and the Epilight energy settings are

roughly 10

> points higher in impact than Photoderm. That is to say that 50j on

Epilight

> equals 60j on Photoderm. The Epilight will handle a 590 filter but

not a

> 550. Not sure about the 570. ***Someone please correct me if I'm

wrong on

> any of the above.***

>

> Here are my settings for tx #5 on 5/26/00:

> Device=Epilight

> Filter=590

> Fluence=52j

> Pulses#=3

> Pulse Duration=3.0ms

> Pulse Delay=20.0ms

>

> Here are my settings for tx #9 on 5/26/00:

> This tx was more aggressive than #5.

> Device=Epilight

> Filter=590

> Fluence=60j

> Pulses#=3

> Pulse Duration=3.0ms

> Pulse Delay=10.0ms

>

> Notice that the joules were higher on tx 9, plus the delay between

pulses

> was shortened from 20 to 10 ms. The shortened pulse delay was

intended to

> cause more heat retention by targeted vessels between pulses. Also

I

> received over 450 blasts (1350+ pulses) during tx 9, as compared to

about

> 320 blasts for tx 5. Please keep in mind that joule settings on

Epilight are

> more aggressive than the Photoderm equivalent. Your doc should know

this but

> you might want to query him if he decides to use the Epilight. Hope

this

> helps.

> ~Jim

>

>

> _________________________________________________________________

> Get your FREE download of MSN Explorer at http://explorer.msn.com

Link to comment
Share on other sites

--- Jim C wrote:

> The 550 filter is shallow penetration and

> concentrates heat at the surface

> layers of the skin, so it generally tends to cause

> more redness and post op

> flareups than the 570 and 590, which penetrate

> deeper and carry heat with

> them.

Sheila--

I did fine after my first PD treatment, which was with

the 570 filter. Txs 2 and 3 utilized the 550, which

made me flare mightily! Strange thing was, once the

flare died down, my skin looked much better than it

had pre-treatment, though it was highly sensitive.

With subsequent txs, I moved back to 570 and 590 with

no further post-tx flaring. My PD clinician told me

that for those with lots of inflammation (evidenced by

things like lots of surface papules), it's quite

common to have flaring after tx with the 550.

__________________________________________________

Get personalized email addresses from Yahoo! Mail - only $35

a year! http://personal.mail.yahoo.com/

Link to comment
Share on other sites

--- Jim C wrote:

> The 550 filter is shallow penetration and

> concentrates heat at the surface

> layers of the skin, so it generally tends to cause

> more redness and post op

> flareups than the 570 and 590, which penetrate

> deeper and carry heat with

> them.

Sheila--

I did fine after my first PD treatment, which was with

the 570 filter. Txs 2 and 3 utilized the 550, which

made me flare mightily! Strange thing was, once the

flare died down, my skin looked much better than it

had pre-treatment, though it was highly sensitive.

With subsequent txs, I moved back to 570 and 590 with

no further post-tx flaring. My PD clinician told me

that for those with lots of inflammation (evidenced by

things like lots of surface papules), it's quite

common to have flaring after tx with the 550.

__________________________________________________

Get personalized email addresses from Yahoo! Mail - only $35

a year! http://personal.mail.yahoo.com/

Link to comment
Share on other sites

--- Jim C wrote:

> The 550 filter is shallow penetration and

> concentrates heat at the surface

> layers of the skin, so it generally tends to cause

> more redness and post op

> flareups than the 570 and 590, which penetrate

> deeper and carry heat with

> them.

Sheila--

I did fine after my first PD treatment, which was with

the 570 filter. Txs 2 and 3 utilized the 550, which

made me flare mightily! Strange thing was, once the

flare died down, my skin looked much better than it

had pre-treatment, though it was highly sensitive.

With subsequent txs, I moved back to 570 and 590 with

no further post-tx flaring. My PD clinician told me

that for those with lots of inflammation (evidenced by

things like lots of surface papules), it's quite

common to have flaring after tx with the 550.

__________________________________________________

Get personalized email addresses from Yahoo! Mail - only $35

a year! http://personal.mail.yahoo.com/

Link to comment
Share on other sites

Join the conversation

You are posting as a guest. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
×
×
  • Create New...