Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 Lew, I have personal experience using an AirAdvice monitor (model 5100). It measures temperature, RH, CO, CO2, particulates and TVOCs. The data collected is uploaded to the AirAdvice website via a phone jack on the side which dials a 800 number. There is no display on the device, the only way to access the data is when you upload it and log on to your account. You can print off raw graphs of the data or use one of their pre- packaged reports. My concern with the device relates to calibration. I'm not so concerned about the temp and RH falling out of calibration as I am the TVOCs and CO2. 's law is that if there's a problem in the building, you won't be there to see the symptoms. " Well everything was fine when I took my measurements. " By doing monitoring and not just point measurements, you better understand how the building is performing. You can measure some of the same parameters with less complex devices (HOBO). The advantage of AirAdvice is the ability to access the data remotely. I have a new Gray Wolf IQ-610 with PID that I will compare with my AirAdvice to see if the numbers come out similar. Ian Cull, PE, CIEC Indoor Sciences, Inc. www.indoorsciences.com > > Everybody, > > I've been aware of this company (AirAdvice) for several years. > Although I have not had personal experience with it's products, it's > product concept always seemed interesting. The company apparently has > a rather large installed base of monitoring equipment and services, > mostly sold, I gather, through AC installers. Now they have put > together some national-level summaries from their monitoring > database, which some may find as interesting as I do. > > Here's the link to see the national summary report (34 months, and a > little over 49,000 structures): > http://airadvice.com/company_info/publications.html > > And a link to see a PDF of a sample report from an individual location: > http://airadvice.com/hvac/about_monitor/AirAdvice_SmartIAQ.pdf > > Does anybody have personal experience with the quality and relevance > of the measurements this system produces? > > Lew Harriman > > ---------------------------------- > Mason-Grant Consulting > P.O. Box 6547 > Portsmouth, NH 03802 > > LewHarriman@... > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 I saw their advertisement and decided to call them to ask questions about their monitor. I left a message on their voice mail over a week ago and I am still waiting for a return call. One question I have is regarding the field calibration of the unit. I have a CO2 data-logging unit that requires calibration before each use; I would expect the same regarding their unit. Does anyone have any information about their unit? I saw the report about their national data. If they don’t field calibrate the unit before each use what good is the data? Bob Air Monitoring Summary - 34 months and 49,000 buildings Everybody, I've been aware of this company (AirAdvice) for several years. Although I have not had personal experience with it's products, it's product concept always seemed interesting. The company apparently has a rather large installed base of monitoring equipment and services, mostly sold, I gather, through AC installers. Now they have put together some national-level summaries from their monitoring database, which some may find as interesting as I do. Here's the link to see the national summary report (34 months, and a little over 49,000 structures): http://airadvice.com/company_info/publications.html And a link to see a PDF of a sample report from an individual location: http://airadvice.com/hvac/about_monitor/AirAdvice_SmartIAQ.pdf Does anybody have personal experience with the quality and relevance of the measurements this system produces? Lew Harriman ---------------------------------- Mason-Grant Consulting P.O. Box 6547 Portsmouth, NH 03802 (603) 431-0635 LewHarrimanMasonGrant -------------------------Confidentiality Notice--------------------------This electronic message transmission contains information from the Los Angeles Department of Water and Power, which may be confidential. If you are not the intended recipient, be aware that any disclosure, copying, distribution or use of the content of this information is prohibited. If you have received this communication in error, please notify us immediately by e-mail and delete the original message and any attachment without reading or saving in any manner. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 Calibration is a very critical issue with these devices. I have had to legal cases where the data was clear bogus. I recently did an IAQ survey comparing 4 instruments that were supposed to be " Calibrated. " All side by side. The worst data was from the CO2 sensor. 2 monitors read 10% low. 1 was 25% too high 1 was 50% too high. Needless to say, use of all of these instruments could lead to the wrong conclusion. I was especially appalled by the monitor that read 50% too high. One could be recommending significant changes to HVAC operation based on erroneous data. Not good. Years ago, it used to be that daily calibration was the norm. Bob Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 Calibration is a very critical issue with these devices. I have had to legal cases where the data was clear bogus. I recently did an IAQ survey comparing 4 instruments that were supposed to be " Calibrated. " All side by side. The worst data was from the CO2 sensor. 2 monitors read 10% low. 1 was 25% too high 1 was 50% too high. Needless to say, use of all of these instruments could lead to the wrong conclusion. I was especially appalled by the monitor that read 50% too high. One could be recommending significant changes to HVAC operation based on erroneous data. Not good. Years ago, it used to be that daily calibration was the norm. Bob Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 I always field check my CO2 instruments with an outdoor reading. Depending on where the building is (how urban), I like to see the outdoor numbers between about 365 and 425. The problem with calibrating these instruments is that there is only one adjustment screw. When you calibrate to about 1,000 ppm CO2 with bottled gas (which is where you want the instrument to be most accurate), the outdoor reading on the very low end of the measurement range can be way off. This can make using the difference between indoor and outdoor readings a joke. Glad you brought it up Ian, Bob and . I'd like to know how others know their CO2 numbers are good at both the low end (350) and high end (1000) when there is only one adjustment screw for calibration. Steve Temes I recently did an IAQ survey comparing 4 instruments that were supposed to be "Calibrated." All side by side. The worst data was from the CO2 sensor. 2 monitors read 10% low. 1 was 25% too high 1 was 50% too high. Needless to say, use of all of these instruments could lead to the wrong conclusion. I was especially appalled by the monitor that read 50% too high. One could be recommending significant changes to HVAC operation based on erroneous data. Not good. Years ago, it used to be that daily calibration was the norm. Bob Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 I always field check my CO2 instruments with an outdoor reading. Depending on where the building is (how urban), I like to see the outdoor numbers between about 365 and 425. The problem with calibrating these instruments is that there is only one adjustment screw. When you calibrate to about 1,000 ppm CO2 with bottled gas (which is where you want the instrument to be most accurate), the outdoor reading on the very low end of the measurement range can be way off. This can make using the difference between indoor and outdoor readings a joke. Glad you brought it up Ian, Bob and . I'd like to know how others know their CO2 numbers are good at both the low end (350) and high end (1000) when there is only one adjustment screw for calibration. Steve Temes I recently did an IAQ survey comparing 4 instruments that were supposed to be "Calibrated." All side by side. The worst data was from the CO2 sensor. 2 monitors read 10% low. 1 was 25% too high 1 was 50% too high. Needless to say, use of all of these instruments could lead to the wrong conclusion. I was especially appalled by the monitor that read 50% too high. One could be recommending significant changes to HVAC operation based on erroneous data. Not good. Years ago, it used to be that daily calibration was the norm. Bob Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 > > Calibration is a very critical issue with these devices. I have had > to legal cases where the data was clear bogus. > > I recently did an IAQ survey comparing 4 instruments that were supposed > to be " Calibrated. " All side by side. > > The worst data was from the CO2 sensor. > > 2 monitors read 10% low. > 1 was 25% too high > 1 was 50% too high. > > Needless to say, use of all of these instruments could lead to the > wrong conclusion. > > I was especially appalled by the monitor that read 50% too high. One > could be recommending significant changes to HVAC operation based on > erroneous data. Not good. > > Years ago, it used to be that daily calibration was the norm. > > Bob > Bob What is an appropriate level of accuracy for measuring CO2? My angle is to look at CO2 levels from intermittent ventilation and in existing systems using lower than recommended amounts of fresh air. Murray Woodgate P.Eng. Arista Engineering Grand Cayman Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 > > Calibration is a very critical issue with these devices. I have had > to legal cases where the data was clear bogus. > > I recently did an IAQ survey comparing 4 instruments that were supposed > to be " Calibrated. " All side by side. > > The worst data was from the CO2 sensor. > > 2 monitors read 10% low. > 1 was 25% too high > 1 was 50% too high. > > Needless to say, use of all of these instruments could lead to the > wrong conclusion. > > I was especially appalled by the monitor that read 50% too high. One > could be recommending significant changes to HVAC operation based on > erroneous data. Not good. > > Years ago, it used to be that daily calibration was the norm. > > Bob > Bob What is an appropriate level of accuracy for measuring CO2? My angle is to look at CO2 levels from intermittent ventilation and in existing systems using lower than recommended amounts of fresh air. Murray Woodgate P.Eng. Arista Engineering Grand Cayman Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 Calibration is a very critical issue with these devices. I have had to legal cases where the data was clear bogus. I recently did an IAQ survey comparing 4 instruments that were supposed to be " Calibrated. " The worst data was from the CO2 sensor. 2 monitors read 10% low. 1 was 25% too high 1 was 50% too high. Needless to say, use of all of these instruments could lead to the wrong conclusion. I was especially appalled by the monitor that read 50% too high. th Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 10, 2007 Report Share Posted July 10, 2007 Calibration is a very critical issue with these devices. I have had to legal cases where the data was clear bogus. I recently did an IAQ survey comparing 4 instruments that were supposed to be " Calibrated. " The worst data was from the CO2 sensor. 2 monitors read 10% low. 1 was 25% too high 1 was 50% too high. Needless to say, use of all of these instruments could lead to the wrong conclusion. I was especially appalled by the monitor that read 50% too high. th Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 12, 2007 Report Share Posted July 12, 2007 I've spoken about the AirAdvice Report at http://airadvice.com/company_info/images/AirAdviceSOIAReport07.pdf with its author, Lucas Klesch, lklesch@... He provided me with the " thresholds " for each parameter (below) that were used to calculate the total number of IAQ problems of each type. These are the levels at which AirAdvice tells its customers that " action is recommended for sensitive groups " (they also have a list of higher/longer exposure thresholds at which the company says action is needed for anyone) Most of the thresholds are quite conservative. The only one I don't like is for CO. AirAdvice reports a problem only if the 8hour avg CO is over 5ppm. I'd prefer if they also reported any level over 9 or at least 35ppm, no matter how brief. I have not yet used the device myself but (CO threshold aside) I like all that I've read about it. --Albert Donnay, MHS AirAdvice thresholds that trigger report of " action recommended to sensitive groups " (or in this report, " IAQ problems " ) ====================================================== Particle Allergens measures .7-10 microns range, over 10ug/m3 Chemical Pollutants (measures total VOCs, calibrated to isobutylene even though not found in indoor air, since it is midrange of C4 to C12) over 500 ug/m3 Carbon Dioxide over 750ppm 24-hour avg (CA school std is 800pppm) Temperature anything above or below ASHRAE normal range of 68 to 75F for one hour or more in winter, or 72 to 78 in summer Humidity anything outside range of 30-55% for one hour or more Carbon Monoxide AirAdvice over 5ppm avg over 8 hours Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 12, 2007 Report Share Posted July 12, 2007 I like the idea AirAdvice has come up with but I have questions about the calibration of the unit. Specifically, how are the CO2 and VOC metering portion of the meter calibrated? Could you please provide me with this information? Thanks Bob MS, CIH, CIEC -----Original Message----- From: iequality [mailto:iequality ]On Behalf Of adonnay@... Sent: Thursday, July 12, 2007 9:37 AM To: iequality Cc: lklesch@...; ADONNAY@... Subject: Re: Air Monitoring Summary - 34 months and 49,000 buildings I've spoken about the AirAdvice Report at http://airadvice.com/company_info/images/AirAdviceSOIAReport07.pdf with its author, Lucas Klesch, lkleschairadvice He provided me with the " thresholds " for each parameter (below) that were used to calculate the total number of IAQ problems of each type. These are the levels at which AirAdvice tells its customers that " action is recommended for sensitive groups " (they also have a list of higher/longer exposure thresholds at which the company says action is needed for anyone) Most of the thresholds are quite conservative. The only one I don't like is for CO. AirAdvice reports a problem only if the 8hour avg CO is over 5ppm. I'd prefer if they also reported any level over 9 or at least 35ppm, no matter how brief. I have not yet used the device myself but (CO threshold aside) I like all that I've read about it. --Albert Donnay, MHS AirAdvice thresholds that trigger report of " action recommended to sensitive groups " (or in this report, " IAQ problems " ) ====================================================== Particle Allergens measures .7-10 microns range, over 10ug/m3 Chemical Pollutants (measures total VOCs, calibrated to isobutylene even though not found in indoor air, since it is midrange of C4 to C12) over 500 ug/m3 Carbon Dioxide over 750ppm 24-hour avg (CA school std is 800pppm) Temperature anything above or below ASHRAE normal range of 68 to 75F for one hour or more in winter, or 72 to 78 in summer Humidity anything outside range of 30-55% for one hour or more Carbon Monoxide AirAdvice over 5ppm avg over 8 hours -------------------------Confidentiality Notice--------------------------This electronic message transmission contains information from the Los Angeles Department of Water and Power, which may be confidential. If you are not the intended recipient, be aware that any disclosure, copying, distribution or use of the content of this information is prohibited. If you have received this communication in error, please notify us immediately by e-mail and delete the original message and any attachment without reading or saving in any manner. Quote Link to comment Share on other sites More sharing options...
Guest guest Posted July 12, 2007 Report Share Posted July 12, 2007 I think you are mistaken and will find the actual CO readings are plotted on a continues time graph. While the graph will show 4ppm as an acceptable limit it will not only plot the actuals but calculate the highest hourly average and highest daily average when the survey is greater than one day. This past January I attended an hour presentation by the sales manager for Airadvice. Then they were not selling their instruments but leasing them three in a package as a sales tool for HVAC dealers and indoor air professionals. The use was to install the monitor for three or more days connected to a telephone line where sometime after midnight the device would upload the days readings for a report to be downloaded at the convenience of the professional. The reason three monitors are leased as a package was so the three could be installed at three different locations within a building and a comparative picture presented for the different areas. For residential surveys the professional would have three units to be surveying three different homes at once allowing a minimum of a three days of readings. If a telephone line is not available the monitor will retain the data in memory where it may be uploaded by the professional after the monitor is retrieved. The monitors if I remember correctly are not able to be calibrated in the field but must be returned to the factory in Portland, Oregon for recalibration. I believe they promise a four day turn around with Fed-Ex. What impressed me was the graphical presentation of the measured levels of particulates and also VOC's over a period of several days. The report introduces the subject of particulates with the statement: " Particle allergens are generally a cause for concern when daily average levels are above 10 ug/m3 " . Then the actual readings are shown plotted with the 10ug/m3 level superimposed. A further discussion of particulates goes on to say: " At levels above 35 ug/m3, they can harm normally healthy adults by causing emphysema and diminished lung capacity " . With that reference the graph invariably will show the home had many hours of particulate contamination in excess of the 35 ug/m3 suggested danger point and the HVAC dealer or IAQ professional is set up to present a program to rectify the problems. A similar reference is given with the VOC report as well as the other reports however it seems the particulate and VOC reports have the greatest impact. Just my two cents..... from last January's seminar. Ken Gibala ============================= > > I've spoken about the AirAdvice Report at > http://airadvice.com/company_info/images/AirAdviceSOIAReport07.pdf > with its author, Lucas Klesch, lklesch@... > > He provided me with the " thresholds " for each parameter > (below) that were used to calculate the total number of IAQ > problems of each type. These are the levels at which > AirAdvice tells its customers that " action is recommended > for sensitive groups " (they also have a list of > higher/longer exposure thresholds at which the company says > action is needed for anyone) > > Most of the thresholds are quite conservative. The only one > I don't like is for CO. AirAdvice reports a problem only > if the 8hour avg CO is over 5ppm. I'd prefer if they also > reported any level over 9 or at least 35ppm, no matter how > brief. > > I have not yet used the device myself but (CO threshold > aside) I like all that I've read about it. > > --Albert Donnay, MHS > > > AirAdvice thresholds that trigger report of " action > recommended to sensitive groups " (or in this report, " IAQ > problems " ) > ====================================================== > Particle Allergens > measures .7-10 microns range, > over 10ug/m3 > > Chemical Pollutants > (measures total VOCs, calibrated to isobutylene even though > not found in indoor air, since it is midrange of C4 to C12) > over 500 ug/m3 > > Carbon Dioxide > over 750ppm 24-hour avg > (CA school std is 800pppm) > > Temperature > anything above or below ASHRAE normal range of 68 to 75F for > one hour or more in winter, or 72 to 78 in summer > > Humidity > anything outside range of 30-55% for one hour or more > > Carbon Monoxide > AirAdvice over 5ppm avg over 8 hours > Quote Link to comment Share on other sites More sharing options...