Subject: AED Recall

<<2007_0307_recall.htm>> For those projects with AED and for those of you

affiliated with organizations (churchs, schools, etc.) that have purchased

AEDs please check to make sure yours isn't one being recalled.

Defibtech Announces a Worldwide Voluntary Recall of Lifeline® and

ReviveR Automated External Defibrillators

FOR IMMEDIATE RELEASE - Guilford, CT -- March 6, 2007 - Defibtech, LLC, is

initiating a voluntary worldwide recall of the Lifeline AED® and ReviveR

AED (semi-automatic external defibrillators). This recall affects all

Lifeline and ReviveR AEDs with software versions 2.002 and earlier. The

self-test software for these devices may allow a self-test to clear a

previously detected low battery condition. If this situation occurs, the

operator may be unaware of the low battery, and the device may be unable to

deliver a defibrillation shock, which could result in failure to resuscitate

a patient.

Defibtech initiated notification of its distributors and customers by letter

on February 22, 2007. Defibtech determined the need for this recall after

learning of three reports of malfunctions from end users. The company has

provided a maintenance procedure that can be used to verify functionality of

the device until the software upgrade has been installed, allowing the device

to remain in service. A copy of this maintenance procedure is being mailed to

customers. This procedure, as well as instructions on determining the

software version of a unit, can be found on the www.defibtech.com/fa2007 web

page. For questions regarding this recall, please refer to the above

referenced web page, contact your distributor or contact Defibtech at

techsupport@..., 1-877-453-4507 or 1-203-453-4507.

Defibtech will provide customers with a free software upgrade to address this

issue for all affected AEDs. This software upgrade will be able to be

installed in the field where the unit is located. The software upgrade is

expected to be available within the next 10 weeks.

Defibtech has distributed approximately 42,000 units worldwide through

distributors to end users including: schools, fire & EMS, businesses, health

clubs and hospitality companies. The products can be identified by the words

" Lifeline AED® " and " ReviveR AED " on the front of the device.

The Food and Drug Administration (FDA) has determined that this action is a

Class I recall. Class I recalls are the most serious type of recall and

involve situations in which there is a reasonable probability that, should

this particular malfunction occur, use of the affected product may cause

serious injury or death if the problem is not corrected. Any adverse

reactions experienced with the use of this product and/or quality problems

should also be reported to the FDA's MedWatch Program by phone at

1-800-FDA-1088, by FAX at 1-800-FDA-0178, by mail at MedWatch, HF-2, FDA,

5600 Fishers Lane, Rockville, MD 20852-9787, or on the MedWatch website at

www.fda.gov/medwatch.

Contacts

Technical Support: 1-877-453-4507, 1-203-453-4507

Media Inquiries: Ray Valek, 708-352-8695, ray@...

------------------------------------------------------------------------

Defibtech <http://www.defibtech.com/index.html>

HOME <http://www.defibtech.com/index.html> SITE MAP

<http://www.defibtech.com/sitemap.html> CONTACT US

<http://www.defibtech.com/contact.html>

Suggest this page to a friend

<http://www.defibtech.com/tellafriend.cgi> EMAIL THIS PAGE

<http://www.defibtech.com/tellafriend.cgi>

Print this page <javascript:print('2007_0129_gul_p.html');> PRINT THIS

PAGE <javascript:print('2007_0129_gul_p.html');>

NEWS

Company News

Home <http://www.defibtech.com/index.html> > News

<http://www.defibtech.com/news/index.html> > Defibtech Announces a

Worldwide Voluntary Recall

*Defibtech Announces a Worldwide Voluntary Recall of Lifeline® and

ReviveR™ Automated External Defibrillators*

FOR IMMEDIATE RELEASE - Guilford, CT -- March 6, 2007 - Defibtech, LLC,

is initiating a voluntary worldwide recall of the Lifeline AED® and

ReviveR AED™ (semi-automatic external defibrillators). This recall

affects all Lifeline and ReviveR AEDs with software versions 2.002 and

earlier. The self-test software for these devices may allow a self-test

to clear a previously detected low battery condition. If this situation

occurs, the operator may be unaware of the low battery, and the device

may be unable to deliver a defibrillation shock, which could result in

failure to resuscitate a patient.

Defibtech initiated notification of its distributors and customers by

letter on February 22, 2007. Defibtech determined the need for this

recall after learning of three reports of malfunctions from end users.

The company has provided a maintenance procedure that can be used to

verify functionality of the device until the software upgrade has been

installed, allowing the device to remain in service. A copy of this

maintenance procedure is being mailed to customers. This procedure, as

well as instructions on determining the software version of a unit, can

be found on the www.defibtech.com/fa2007

<http://www.defibtech.com/fa2007> web page. For questions regarding this

recall, please refer to the above referenced web page, contact your

distributor or contact Defibtech at techsupport@...

<mailto:techsupport@...>, 1-877-453-4507 or 1-203-453-4507.

Defibtech will provide customers with a free software upgrade to address

this issue for all affected AEDs. This software upgrade will be able to

be installed in the field where the unit is located. The software

upgrade is expected to be available within the next 10 weeks.

Defibtech has distributed approximately 42,000 units worldwide through

distributors to end users including: schools, fire & EMS, businesses,

health clubs and hospitality companies. The products can be identified

by the words " Lifeline AED® " and " ReviveR AED™ " on the front of the device.

The Food and Drug Administration (FDA) has determined that this action

is a Class I recall. Class I recalls are the most serious type of recall

and involve situations in which there is a reasonable probability that,

should this particular malfunction occur, use of the affected product

may cause serious injury or death if the problem is not corrected. Any

adverse reactions experienced with the use of this product and/or

quality problems should also be reported to the FDA's MedWatch Program

by phone at 1-800-FDA-1088, by FAX at 1-800-FDA-0178, by mail at

MedWatch, HF-2, FDA, 5600 Fishers Lane, Rockville, MD 20852-9787, or on

the MedWatch website at www.fda.gov/medwatch <http://www.fda.gov/medwatch>.

*Contacts*

Technical Support: 1-877-453-4507, 1-203-453-4507

Media Inquiries: Ray Valek, 708-352-8695, ray@...

<mailto:ray@...>

*PHONE: 1-866-DEFIB-4-U (1-866-333-4248)

EMAIL: sales@... <mailto:sales@...> * *HOME

<http://www.defibtech.com/index.html> PRODUCTS

<http://www.defibtech.com/products/> TECHNOLOGY

<http://www.defibtech.com/technology.html> NEWS

<http://www.defibtech.com/news/> RESOURCES

<http://www.defibtech.com/resources.html> DOWNLOADS

<http://www.defibtech.com/downloads.html> ABOUT

<http://www.defibtech.com/aboutus.html> CONTACT

<http://www.defibtech.com/contact.html>*

Link to comment
Share on other sites

Join the conversation

You are posting as a guest. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
×
×
  • Create New...