Guest guest Posted April 5, 2010 Report Share Posted April 5, 2010 <snip> would you be interested in a head to foot type of godzilla using a low voltage and slowly adjusting upwards till we get some sort of effect? I could try to design that but I don't want to do the work unless it's going to have a person who will test and report. The polarity reversals WILL make it unusable as described. Your trouble will be lard on your back & belly that has some times 30 the normal capacitance plus about times 10 the bulk resistance due to the poor blood supply [capillaries only]. Have found the limit of tolerance to be about 2 mA into at least 10 square inches [each] and that's with a steady state of 2.85 Vdc across the pads, CONSTANT CURRENT, BIDIRECTIONAL FILTRATION. To improve the treatment - have designed [but not yet tested] a gradually increasing / decreasing ramp waveform with about a 6 second duration to gradually let the polarity switch - besides the built in 6 second delay from one lead being shorted to the other. On a meter, without the constant current filter - would take over 30 seconds total to do the reversal - with under 1 mA / 10 sq. " pads, BITE EVERY TIME. If you were bulimic - might put up with it. Wrist / ankle tissues don't even need a filter for up to 2 mA into a 10 sq " pad pair; with a 6 second delay that shorts the first pad to the second pad voltage; never feel a thing during reversals. Joe. Quote Link to comment Share on other sites More sharing options...