Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 When I had the infected puncture wound on my foot I applied the sticky conductors each side of the wound and that did the job in less than 24 hours. Jack Re: Re: protocol ?? for non-healing wound on a diabetic? thanks for the input. Now my question would be where the placement would be to heal the wound. Would it be above or below the wound? could you use acupresure points? etc etc etc baby_grand <bobluhrs@...> wrote: no! No! No! Never put electrodes on broken skin. They are transdermal only. Never use bare metal electrodes, either, they should be wrapped with cloth, always. One fellow killed himself sticking needles into the body and applying very low voltage, figuring it was " safe " . It wasn't. Even a AAA battery used that way can conceivably kill you, since the resistance drops way down without the skin and the clot to provide some resistance and limit the current. Just use it on skin, not on raw wounds!!! Please? OK! bG > > > > " " protocol ?? for non-healing wound on a diabetic? " " > > This came up on the oxyplus list and naturally they talk about > ozone. > > But shouldn't godzila also work ?? and if so what would be a > treatment to > > try. > > Thanks for to all you wonderful people for being there. > > ivan in SF bay area > > > > > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 V, According to one source online: " The electro-acupuncture device is not intended to provide a significant current between the acupuncture needles: it delivers less than a milliampere (modern devices usually have an upper level of 0.6 milliampere), about the same as that produced by a wristwatch battery. But, it will provide a significant voltage: 40-80 volts (with short spikes as high as 130 volts), which is the basis for the patient response. In the commonly-used portable battery devices, this is accomplished by boosting the voltage output of the battery, such as raising the voltage from 9 volts to 45 volts. Thus, there is virtually no current transmitted through the body, but there is enough of a local voltage stimulus for the patient to feel an evident reaction at the point where the needle is inserted. " It seems to me that .6 of a milliampere is comparable to a beck device's output. Am I wrong? There is another online journal article describing " Effect of electro-acupuncture stimulation of different frequencies and intensities on ovarian blood flow in anaesthetized rats " . In this experiment, three different intensities were applied to the needles of - 1.5, 3, and 6 mA - were applied for 35 s. Aren't these levels way beyond what is used for Godzilla or am I missing something? None of these rats were electrocuted. doug Re: Re: protocol ?? for non-healing wound on a diabetic? Hi d.a., The power is very low on accupuncture needles. What we use here with batteries is much higher and not safe for the needles. Take care, V Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 distinguish between a battery and a device with a limiting resistor in it that keeps the current down. a battery like our 6v one, has no such limitations, and the current depends only on the resistance in the body. If you cut the skin, the resistance drops hundreds of times. the device would still output the same current, or close to it, since it has built-in current limiters, but the battery has only its own internal resistance, which is quite low. thus, the body would accept tons of current all of a sudden from a battery, but the same or close, from a controlled device. you will observe that here we are using simple batteries to do the work, figuring that people will have more access to this if it's kept simple. but...there's this factor that you cannot use batteries on open wounds or via needles, etc. It's a severe warning given here. bG > > V, > > According to one source online: > > " The electro-acupuncture device is not intended to provide a significant current between the acupuncture needles: it delivers less than a milliampere (modern devices usually have an upper level of 0.6 milliampere), about the same as that produced by a wristwatch battery. But, it will provide a significant voltage: 40-80 volts (with short spikes as high as 130 volts), which is the basis for the patient response. In the commonly-used portable battery devices, this is accomplished by boosting the voltage output of the battery, such as raising the voltage from 9 volts to 45 volts. Thus, there is virtually no current transmitted through the body, but there is enough of a local voltage stimulus for the patient to feel an evident reaction at the point where the needle is inserted. " > > It seems to me that .6 of a milliampere is comparable to a beck device's output. Am I wrong? > > There is another online journal article describing " Effect of electro-acupuncture stimulation of different frequencies and intensities on ovarian blood flow in anaesthetized rats " . In this experiment, three different intensities were applied to the needles of - 1.5, 3, and 6 mA - were applied for 35 s. Aren't these levels way beyond what is used for Godzilla or am I missing something? None of these rats were electrocuted. > > doug > Re: Re: protocol ?? for non- healing wound on a diabetic? > > > Hi d.a., > > The power is very low on accupuncture needles. What we use here with batteries is much higher and not safe for the needles. > > > > > Take care, > V > > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 Ok, I think, I understand now. A simple battery device as you are promoting on this list can be deadly with broken skin, but a more sophisticated device with limiting resistors can be used safely. At one time, I was playing with the notion of connecting V's Godzilla device to acupuncture needles for animal veterinary experimentation. Is such use of this device, a cheap alternative to the more expensive electro-acupuncture devices out there---in theory, anyway? thanks, doug Re: protocol ?? for non-healing wound on a diabetic? distinguish between a battery and a device with a limiting resistor in it that keeps the current down. a battery like our 6v one, has no such limitations, and the current depends only on the resistance in the body. If you cut the skin, the resistance drops hundreds of times. the device would still output the same current, or close to it, since it has built-in current limiters, but the battery has only its own internal resistance, which is quite low. thus, the body would accept tons of current all of a sudden from a battery, but the same or close, from a controlled device. you will observe that here we are using simple batteries to do the work, figuring that people will have more access to this if it's kept simple. but...there's this factor that you cannot use batteries on open wounds or via needles, etc. It's a severe warning given here. bG Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 probably not. they work by stimulation, and that requires the right type of electricity waveform to actuate the nerves and whatever else it does. Stick with the pro devices if you are entering that realm. bG > > Ok, I think, I understand now. A simple battery device as you are promoting on this list can be deadly with broken skin, but a more sophisticated device with limiting resistors can be used safely. At one time, I was playing with the notion of connecting V's Godzilla device to acupuncture needles for animal veterinary experimentation. Is such use of this device, a cheap alternative to the more expensive electro-acupuncture devices out there---in theory, anyway? > > thanks, > > doug > Re: protocol ?? for non- healing wound on a diabetic? > > > > distinguish between a battery and a device with a limiting resistor > in it that keeps the current down. a battery like our 6v one, has no > such limitations, and the current depends only on the resistance in > the body. If you cut the skin, the resistance drops hundreds of > times. the device would still output the same current, or close to > it, since it has built-in current limiters, but the battery has only > its own internal resistance, which is quite low. thus, the body > would accept tons of current all of a sudden from a battery, but the > same or close, from a controlled device. > > you will observe that here we are using simple batteries to do the > work, figuring that people will have more access to this if it's kept > simple. but...there's this factor that you cannot use batteries on > open wounds or via needles, etc. It's a severe warning given here. > > bG > > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 Hi d.a., Well the key point there being the short spikes . you can have the cerruntts that high with very short spikes. but try that same trick with a continuos currentn like from a battery then you got a big problem. When they are using frequencies and short spikes the current is not an issue becaues the durlation is so short. Whith our devices there is no current timing regulation and that then provides too much. So yes the current is the same as a beck device but the duration and timing is very different so you dont want to mix those up and think that runnnig the same amount of current through a needle without the proper timing and regulation is gonig to be safe. It is not. For examplre my Electro mucsle stimulator runs on a 9 volt battery and it is stepped up to about 360 volts but is very short voltage spikes (1000th of a second) and it is quite bearable and makes the muscles jerk. take the same 360 volts and appyl it for a long durantion of even a full second and it colud kill you if you have the muscel pads in the wrong place. so when worknig with teh electricty there are many factors to consider beyond the current when applying to the body, especially when you bypass the skin which has a high resistance, then you quickly run into dangerous territory and it requires very precise application of the current. you gotta know exactly what your doing there Take care, V > V, > According to one source online: > " The electro-acupuncture device is not intended to provide a significant > current between the acupuncture needles: it delivers less than a > milliampere (modern devices usually have an upper level of 0.6 > milliampere), about the same as that produced by a wristwatch battery. > But, it will provide a significant voltage: 40-80 volts (with short > spikes as high as 130 volts), which is the basis for the patient > response. In the commonly-used portable battery devices, this is > accomplished by boosting the voltage output of the battery, such as > raising the voltage from 9 volts to 45 volts. Thus, there is virtually no > current transmitted through the body, but there is enough of a local > voltage stimulus for the patient to feel an evident reaction at the point where the needle is inserted. " > It seems to me that .6 of a milliampere is comparable to a beck device's output. Am I wrong? > There is another online journal article describing " Effect of > electro-acupuncture stimulation of different frequencies and intensities > on ovarian blood flow in anaesthetized rats " . In this experiment, three > different intensities were applied to the needles of - 1.5, 3, and 6 mA - > were applied for 35 s. Aren't these levels way beyond what is used for > Godzilla or am I missing something? None of these rats were electrocuted. > doug > Re: Re: protocol ?? for non-healing wound on a diabetic? > Hi d.a., > The power is very low on accupuncture needles. What we use here with > batteries is much higher and not safe for the needles. > Take care, > V > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 right,could not agree more. the whole idea of muscle stimulators rests on a reliably controlled duration. If the device breaks down and does not cut out after the 1000/th second or so, you could have troubles with it. that's highly unlikely. I often wondered about that, too, then I saw the waveform plotted out vs time, and it was definitely near-instant, and not many electrons (current) flows into you in that small a time period. electrons cause ionization, create toxins, etc in large amounts. but a short spike supplies very few of them. if the spike is high-intensity it affects the nerves as the body tries to react to avoid it. but there's few significant chemical changes that could kill you going on. we are deliberately creating these chemical changes at a controlled level and avoiding the breakdown of healthy tissue, and nerve stimulus that could be a danger, and instead inactivating microbes by this tiny amount of chemical change that affects " them " and not " us " . bG > > Hi d.a., > > Well the key point there being the short spikes . you can have the cerruntts that high with very short spikes. but try that same trick with a continuos currentn like from a battery then you got a big problem. When they are using frequencies and short spikes the current is not an issue becaues the durlation is so short. Whith our devices there is no current timing regulation and that then provides too much. So yes the current is the same as a beck device but the duration and timing is very different so you dont want to mix those up and think that runnnig the same amount of current through a needle without the proper timing and regulation is gonig to be safe. It is not. > For examplre my Electro mucsle stimulator runs on a 9 volt battery and it is stepped up to about 360 volts but is very short voltage spikes (1000th of a second) and it is quite bearable and makes the muscles jerk. take the same 360 volts and appyl it for a long durantion of even a full second and it colud kill you if you have the muscel pads in the wrong place. so when worknig with teh electricty there are many factors to consider beyond the current when applying to the body, especially when you bypass the skin which has a high resistance, then you quickly run into dangerous territory and it requires very precise application of the current. you gotta know exactly what your doing there > > > > > Take care, > V > > > > V, > > > According to one source online: > > > " The electro-acupuncture device is not intended to provide a significant > > current between the acupuncture needles: it delivers less than a > > milliampere (modern devices usually have an upper level of 0.6 > > milliampere), about the same as that produced by a wristwatch battery. > > But, it will provide a significant voltage: 40-80 volts (with short > > spikes as high as 130 volts), which is the basis for the patient > > response. In the commonly-used portable battery devices, this is > > accomplished by boosting the voltage output of the battery, such as > > raising the voltage from 9 volts to 45 volts. Thus, there is virtually no > > current transmitted through the body, but there is enough of a local > > voltage stimulus for the patient to feel an evident reaction at the point where the needle is inserted. " > > > It seems to me that .6 of a milliampere is comparable to a beck device's output. Am I wrong? > > > There is another online journal article describing " Effect of > > electro-acupuncture stimulation of different frequencies and intensities > > on ovarian blood flow in anaesthetized rats " . In this experiment, three > > different intensities were applied to the needles of - 1.5, 3, and 6 mA - > > were applied for 35 s. Aren't these levels way beyond what is used for > > Godzilla or am I missing something? None of these rats were electrocuted. > > > doug > > Re: Re: protocol ?? for non-healing wound on a diabetic? > > > > Hi d.a., > > > The power is very low on accupuncture needles. What we use here with > > batteries is much higher and not safe for the needles. > > > > > > Take care, > > V > > > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 Careful, Doug. I am old enough to remember the days when you could drown extra, unwanted pups or kittens, but today you can go to jail for that. Its called animal cruelty. Its possible that you could call down the wrath of the system on yourself if they discover that you have electrocuted your dog. Dick Re: protocol ?? for non-healing > wound on a diabetic? > > > > distinguish between a battery and a device with a limiting resistor > in it that keeps the current down. a battery like our 6v one, has no > such limitations, and the current depends only on the resistance in > the body. If you cut the skin, the resistance drops hundreds of > times. the device would still output the same current, or close to > it, since it has built-in current limiters, but the battery has only > its own internal resistance, which is quite low. thus, the body > would accept tons of current all of a sudden from a battery, but the > same or close, from a controlled device. > > you will observe that here we are using simple batteries to do the > work, figuring that people will have more access to this if it's kept > simple. but...there's this factor that you cannot use batteries on > open wounds or via needles, etc. It's a severe warning given here. > > bG > > > Quote Link to comment Share on other sites More sharing options...
Guest guest Posted December 30, 2005 Report Share Posted December 30, 2005 right. You need the milliamp, smaller meter, Loew's $11.95. test it for continuity before you leave the store, just rip it open, set on the bottom-most setting (continuity check) and touch probes together. Abt 1/3 of these meters are dead, so check it first. b > > > > > > > > " " protocol ?? for non-healing wound on a diabetic? " " > > > > This came up on the oxyplus list and naturally they talk about > > > ozone. > > > > But shouldn't godzila also work ?? and if so what would be a > > > treatment to > > > > try. > > > > Thanks for to all you wonderful people for being there. > > > > ivan in SF bay area > > > > > > > > > > > > Quote Link to comment Share on other sites More sharing options...